nameylf
第34楼2007/11/06
上传一份美国Spellman公司高压电源的资料,里面详细列举了作为XRF高压电源,应该提供的,或者说,应用于XRF仪器时,高压电源的一些技术要求。
美国Spellman公司高压电源资料
nameylf
第36楼2007/11/07
SPELLMAN HIGH VOLTAGE ELECTRONICS CORPORATION
• 50KV AT 2 MA. 50 WATTS MAX.
• ADJUSTABLE INTEGRATED FILAMENT SUPPLY
• OVERVOLTAGE & SHORT CIRCUIT PROTECTION
• VOLTAGE & CURRENT PROGRAMMING
• LOCAL AND REMOTE EMISSION CONTROL
• SAFETY INTERLOCK
• OEM CUSTOMIZATION AVAILABLE
• CE MARKED
• UL RECOGNIZED
The MNX Series is the result of Spellman’s exceptional
high voltage packaging and surface mount fabrication
techniques, coupled with its proprietary encapsulation
technology producing this ultra compact-sized OEM 50
Watt X-ray generator module.
The MNX Series is designed to power grounded cathode
X-ray tubes from a variety of well known manufacturers. It
features a 0 to 50kV high voltage output, and up to 2mA
of emission current limited to 50 Watts, operating from a
+24Vdc input. The MNX utilizes a closed loop filamentry
beam control circuit to provide a highly regulated beam
current. The ground referenced low noise dc filament
supply operates between 0.3 and 3.5 amps. Offering tight
regulation, high stability and low ripple, the MNX provides
users both local and remote analog control to set beam
voltage, emission current and filament current limit. An
optional USB, RS232 or ethernet interface is available.
TYPICAL APPLICATIONS
Powering grounded cathode X-ray tubes from
Kevex, Oxford, RTW, Superior, Varian and Trufocus.
OPTIONS
XCC XRM Compatible HV Cable
SIC Standard Interface Controller
(Ethenet, USB & RS232)
5VPM 0 to 5 Volt Programming and Monitor Scaling
SPECIFICATIONS
Input:
+24Vdc±10%, 4.0A maximum.
Output:
0 to 50 kV at 0 to 2 mA, limited to a maximum of 50 watts.
Voltage Control:
Local: Internal multi-turn potentiometer to set voltage
from 0 to full output voltage.
Remote: 0 to +10Vdc proportional from 0 to full output voltage.
Accuracy: ±1%. ZIN: 10Mohm.
Emission Control:
Local: Internal potentiometer to set beam current
between 0 and full output current.
Remote: 0 to +10Vdc proportional from 0 to full output current.
Accuracy : ±1%. ZIN: 10Mohm.
Filament limit and filament preheat control
capability is also provided.
DC Filament Supply:
Current: 3.5A, adjustable limit
Voltage: 5.0 volt limit
Voltage Regulation:
Load: 0.01% of output voltage no load to full load.
Line: ±0.01% for ±10% change in input voltage.
Current Regulation:
Load: 0.01% of output current from 0 to rated voltage.
Ripple:
0.1% p-p of maximum rated output voltage.
Environmental:
Operational: 0°C to +50°C
Storage: -40°C to +85°C
Humidity: 0% to 90%, non-condensing
Temperature Coefficient:
0.01% per °C, voltage and current.
Stability:
0.05% per 8 hours after 1/2 hour warm-up.
Voltage and Current Monitors:
0 to +10Vdc proportional from 0 to rated output.
Accuracy ±1%.
Dimensions:
Standard Unit: 5.00”H x 2.87”W x 8”D
(127.00mm x 72.90mm x 203.25mm).
XCC Option: 5.00”H x 2.87”W x 9”D
(127.00mm x 72.90mm x 228.65mm).
SIC Option: 5.75”H x 2.87”W x 8”D
(146.05mm x 72.90mm x 203.25mm).
Weight:
6.5 lbs. (2.9kg)