一道黑
第11楼2008/11/01
VIII. STATISTICAL ANALYSIS
A. General
Methods validation includes an assessment of the adequacy of the analytical procedure. Statistical analysis (e.g., linear regression analysis, relative standard deviation) of methods validation data is often used to demonstrate the validity of the method. The statistical procedures for the analysis of the validation data should be determined prior to the start of any validation study. The procedure followed, including the amount of data to collect and the criteria used in determining the acceptability of the analytical procedure, should be specified.
The raw methods validation data and statistical procedures used to analyze the raw data should be provided and discussed in the sections on analytical procedures and controls. All statistical procedures used in the analysis of the data should be based on sound principles and be suitable for evaluating the dataset.
VIII. 统计分析
A.基本原则
(比如:线性回归分析,相对标准偏差)以说明方法的正确性。在开始分析方法验证之前,应当就要确定用于验证资料分析的统计方法。还应当要规定所要遵循的程序,包括所需采集的数据量和确定分析方法合适性的合格标准。
应当要在分析方法和控制章节中提供和讨论分析方法验证原始资料和所用的统计方法。所有用于数据分析的统计程序都应当是科学的,并适用于评估该数据群的。
B. Comparative Studies
Comparative studies are performed to evaluate intermediate precision (e.g., different equipment, analysts, days). Comparative studies are also used to evaluate between laboratory variability (i.e., reproducibility) when an analytical procedure is used in more than one laboratory or to compare and evaluate the precision and accuracy of two analytical procedures (e.g., regulatory analytical procedure and an alternative analytical procedure). When comparative studies are performed, homogeneous samples from the same batch should be used, if feasible. Comparative results should be statistically analyzed and discussed and any bias explained.
B:对比研究
开展对比研究以评估中间精密度(如,不同设备,不同分析员,不同天等)。当一分析方法会在多个实验室应用时,或要比较和评估两个分析方法(比如,法定分析方法和替代分析方法)的精密度和准确度时,也会进行对比研究以评估实验室间的差异(也就是,重现性)。在进行对比研究时,应当要尽可能地使用同一批号的均匀样品。需对对比研究的结果进行统计分析和讨论,并对偏差进行解释。
C. Statistics
For information on statistical techniques used in making comparisons, as well as other general information on the interpretation and treatment of analytical data, appropriate literature or texts should be consulted (see references) .
C:统计
关于用于对比分析的统计技术资料,和用于分析数据处理和解析的其它基本资料,可参见相关的文献(见参考文献)。
一道黑
第12楼2008/11/01
IX. REVALIDATION
When sponsors make changes in the analytical procedure, drug substance (e.g., route of synthesis), or drug product (e.g., composition), the changes may necessitate revalidation of the analytical procedures. Revalidation should be performed to ensure that the analytical procedure maintains its characteristics (e.g., specificity) and to demonstrate that the analytical procedure continues to ensure the identity, strength, quality, purity, and potency of the drug substance and drug product, and the bioavailability of the drug product. The degree of revalidation depends on the nature of the change. When a different regulatory analytical procedure is substituted (e.g., HPLC for titration), the new procedure should be validated (see section VII).
IX. 再验证
当发起人对分析方法,原料药(比如,合成路线),或制剂(比如,组分)作了更改的话,则需要对分析方法进行重验证。进行重验证是为了确保该分析方法仍然保持其特性(比如,专属性),并论证说明该分析方法仍然能确保原料药和制剂的同一性,浓度/剂量,质量,纯度和功效,及制剂的生物利用度。重验证的程序取决于该变更的性质。当使用了另一个法定分析程序的话(比如,用HPLC代替了滴定法),则新的分析方法也需要验证(见第VII章)。
If during each use an analytical procedure can meet the established system suitability requirements only with repeated adjustments to the operating conditions stated in the analytical procedure, the analytical procedure should be reevaluated, amended, and revalidated, as appropriate.
FDA intends to provide guidance in the future on postapproval changes in analytical procedures.
如果在每次使用时,都必须要对分析方法中所述的操作条件进行反复调整,才能使其符合系统适应性要求的话,则该分析方法需要适当进行重新评估,修正和重验证。
一道黑
第13楼2008/11/01
X. METHODS VALIDATION PACKAGE: CONTENTS AND PROCESSING
Part of the methods validation process may include FDA laboratory analysis to demonstrate that an analytical procedure is reproducible by laboratory testing. A methods validation package (see X.A) and samples (see X.B) will be needed for this process.
A. Methods Validation Package
The methods validation package will usually include information copied from pertinent sections of the application. To aid the review chemist, these copies should retain the original pagination of the application sections. For ANDA and NDA products, the archival copy and extra copies of the methods validation packages should be submitted with the application. For ANDAs and related supplemental applications, one archival copy and two extra copies of the methods validation package should be submitted. For NDAs and related supplemental applications, one archival copy and three extra copies should be submitted. For BLAs and PLAs, a separate methods validation package need not be submitted. Information similar to that specified here should be included in the BLA or PLA submission.
X. 分析方法验证资料:内容和数据处理
FDA实验室的分析以论证说明某一分析方法是能被重现的。在这个过程中将会需要分析方法验证资料(见X.A)和样品(见X.B)。
A.分析方法验证资料
分析方法验证资料通常会包括申请中的相关章节。为了便于评审化学家进行评审,这些资料应当要和其在原来申请中一样,包括内容和形式(archival copy)和其它副本。对于仿制药申请和其相关的补充申请,需要提交一份分析方法验证资料的存档副本(archival copy)和另外两份副本。对于新药申请及其相关补充申请,需要提交一份分析方法验证资料的存档副本(archival copy)和另外三份副本。对于BLA和PLA,则不需要单独递交分析方法验证资料。类似的资料应当摆在BLA和PLA申请中。
The methods validation package should include:
1. Tabular List of All Samples to Be Submitted
The list should include the lot number, identity (with chemical name and structure
where required for clarity), package type and size, date of manufacture, and quantity of the samples.
2. Analytical Procedures
A detailed description of each of the analytical procedures listed in the specifications should be submitted. The description should be sufficient to allow the FDA laboratory analysts to perform the analytical procedure (see section VI).
3. Validation Data
Appropriate validation data to support the analytical procedures should be submitted. Individual values as well as summary tables should be provided. Representative instrument output and raw data and information regarding stress studies should be included (see section VII).
4. Results
The results obtained by the applicant for the submitted samples should be provided. Alternatively, COAs could be submitted. The dates of analysis should be stated.
5. Composition
The components and composition of the drug product should be provided.
6. Specifications
The specifications for the drug substance and the drug product should be included.
7. Material Safety Data Sheets
The applicant should include material safety data sheets (MSDSs) for all samples, standards, and reagents (29 CFR 1910.1200(g)). As appropriate, MSDSs should be provided for other materials used in the analytical procedures listed in the methods validation package. In the case of toxic or hazardous materials, MSDSs should be posted on the outside of the package to facilitate safe handling.
分析方法验证资料应当要包括:
1.所需递交样品的列表清单
(化学名和结构式),包装类型和大小,生产日期,样品量。
2.分析方法
FDA实验室分析人员根据这个描述进行操作。(见第VI章)
3.验证资料
(见第VII章)。
4.结果
应当要提供申请者对所提供样品所做分析的分析结果,或者提供其相应的分析报告单。需说明分析日期。
5.组分
需说明制剂的组分和组成。
6.质量标准
需提供原料药和制剂的质量标准。
7.安全数据表
申请者应当要提供所有样品,标准品和试剂的安全数据表(MSDS)(29CFR 1910.1200(g))。还要适当提供分析方法验证中所列各分析方法所有的其它物料的安全数据表(MSDS)。如果是毒性物料或危险性物料,则在外包装上要贴上MSDS,以便于安全处理。
B. Selection and Shipment of Samples
On request from CDER, an NDA or ANDA applicant must submit samples of drug product, drug substance, noncompendial reference standards, and blanks, so that the suitability of the applicant=s drug substance and drug product analytical procedures can be evaluated by FDA laboratories (21 CFR 314.50(e) and 314.94(a)(10)). For BLAs and PLAs, representative samples of the product must be submitted, and summaries of the results of tests performed on the lots represented by the submitted sample must be provided (21 CFR 601.2(a) and 601.2(c)(1)(vi)).
一道黑
第14楼2008/11/01
B:样品的选择和运输
NDA或ANDA申请者必须要根据药品评审和研究中心(CDER)的要求递交制剂,原料药,非药典对照品和空白,以使FDA实验室可以评估申请者所用制剂和原料药分析方法的适用性。(21CFR 314.50(e) 和314.94(a)(10))。对于BLA和PLA,需提交产品的代表性样品,并提供所提交样品批次的检测结果。(21CFR601.2(a)和601.2(c)(1)(vi)).
For CDER products, the number of sets of samples that should be submitted for methods validation will be identified in the instructions forwarded to the applicant by the FDA laboratory. In general, the quantity of samples in each set should be double the amount needed to carry out the testing as performed by the applicant. Along with the drug substance and the drug product samples, the applicant should submit internal standards, non-USP reference standards, samples of impurities, degradation products, and unusual reagents. A set of samples will be shipped to each assigned laboratory.
对于CDER产品,FDA实验室会告诉申请者所需递交样品的量。一般来说,样品量应当是实验用量的两倍。除了递交原料药和制剂样品之外,申请者还应当要递交内部对照品,非美国药典对照品,杂质样品,降解物和非常用试剂。应当要向每个指定的实验室寄送一系列样品。
For biological products, CBER should be consulted on the submission of samples and supporting materials.
Unless specified differently by the reviewer, samples from any batch, preferably samples from an aged batch, may be selected for NDAs and NDA supplemental applications. The submitted drug product samples should be from a batch made with the proposed market formulation. For ANDAs and appropriate supplements, a sample of the finished product from a batch being used to support approval of the submission should be used. If a sample is selected from a batch not described in the application, an amendment containing a copy of the batch record and certificate of analysis should be provided to the ANDA. For supplements that do not require submission and review of an exhibit batch record and associated data, any commercial batch may be submitted. For biological products, samples from several consecutively manufactured batches should be submitted.
CBER咨询关于样品和支持资料的递交。
除非评审官另有说明,任一批次的样品,最好是较早批次,都可以用于新药申请及其补充申请。所递交的制剂样品必须是根据拟定的上市配方生产的。对于仿制药申请及其相关的补充,应当要提交用于支持申请批准的制剂批次的样品。如果样品是来自于一申请中未提及的批次的话,则在ANDA中还应当要补充一份批记录和分析报告单的复印件。对于不需要递交申请和不需要审阅批记录及相关资料的补充,可递交任一商业批次的样品。对于生物制品,应当要提交连续几个生产批次的样品。
The drug product should be supplied in its original packaging. Bulk substances (e.g., drug substances, impurities, excipients) should be stored in opaque nonreactive containers. To prevent breakage during shipping, the samples should be adequately packaged in a sturdy container. Samples shipped from outside the United States should contain the appropriate customs forms to reduce delay in delivery.
If special storage precautions (e.g., freezing, use of an inert gas blanket) are required to protect sample integrity, arrangements should be made in advance with the validating laboratory for scheduled direct delivery. If a sample is toxic or potentially hazardous, the container should be prominently labeled with an appropriate warning and precautionary handling instructions.
应当要以其原包装提供制剂样品。而像原料药,杂质,赋形剂等,则应当要保存在不透明的惰性容器中。为了防止在运输过程中泄露,样品应当要装在耐用的容器中。如果是美国国外的样品,则应当要有适当的海关单据,以减少耽搁。
如果样品需要特殊的储存条件(比如,冷冻,惰性气体保护),则要事先和验证实验室联系以安排直接递送。如果是毒性样品或危险性样品,则在容器的显著位置标明警示标志和预防措施。
一道黑
第15楼2008/11/01
C. Responsibilities of the Various Parties
1. Applicant
In the sections of the application on analytical procedures and controls, the applicant should provide a name, address, telephone number, and facsimile number so that samples can be requested. If this information is not provided, the contact person and address listed in the NDA, ANDA, BLA, or PLA submission will be used.
The methods validation packages should be compiled and submitted with the NDA or ANDA submission. For BLAs and PLAs, a separate methods validation package need not be submitted. When an FDA laboratory contacts the applicant for samples, the applicant should provide FDA laboratories with the samples within 10 working days. With the exception of sample delivery arrangements, all communications concerning validation at the FDA laboratories should be made through or with the knowledge of the review chemist for CDER applications, or the BLA/PLA committee chair for CBER applications.
C:各方职责
1.申请人
FDA可以向其发送提交样品的要求。如果没有提供这些信息的话,则会用NDA,ANDA,BLA或PLA申请中所写地址和联系人信息进行联系。
NDA或ANDA资料一起编写和递交。对于BLA和PLA,则不需要递交单独的分析方法验证资料。
一旦FDA实验室要求申请人提交样品的话,申请人应当要在10个工作日将样品提供给FDA实验室。除了样品运送之外,所有关于在FDA实验室进行验证的交流工作都要当要让CDER申请的化学评审官知道,如果是CBER申请,则要让BLA/PLA委员会主席知道。
2. Review Chemist
The review chemist will review the application to determine that the analytical procedures are adequate to ensure the identity, strength, quality, purity, and potency of the drug substance and/or drug product. Any changes in the methods resulting from the review of the application may require resubmission of the methods validation package. The review chemist, in coordination with the appropriate FDA laboratories, will decide which analytical procedures are to be validated. Comments from the FDA laboratories, if any, will be forwarded by the review chemist to the applicant on completion of the studies by the laboratories.
2.化学评审官
化学评审官负责审核分析方法验证报告(包括鉴别,浓度/剂量,质量,纯度和功效。对申请进行评审后,如要对分析方法作必要修改的话,则需要重新递交分析方法验证资料。化学评审官会和相关的FDA实验室进行讨论,确定需要对哪个分析方法进行验证。化学评审官会将FDA实验室对所做研究的意见转给申请人。
3. FDA Laboratory
An FDA laboratory will contact applicants with instructions on the submission of samples and the addresses to which samples should be mailed. The laboratory will test the samples according to the submitted analytical procedures to determine whether the analytical procedures are acceptable for quality control and suitable for regulatory purposes. Results and comments will be forwarded to the review chemist on completion of the studies.
4. Investigator
The investigator inspects the analytical laboratory testing sites where the release and stability testing are performed to ensure that the analytical procedures are performed in compliance with CGMP/GLP.
3.FDA实验室
FDA实验室会和申请者联系,告知样品递交程序和注意事项及邮寄地址。FDA实验室会所递交的分析方法对样品进行检测,以确定该分析方法是否适用于质量控制,并符合法规要求。在完成研究之后,FDA实验室会将结果和意见转给化学评审官。
4.检查官
检查官会对进行放行检测和稳定性实验的分析实验室进行检查,以确保所做的分析检测能符合CGMP/GLP。
一道黑
第16楼2008/11/01
XI. METHODOLOGY
Sections II through IX provide general information on the submission of analytical procedures and methods validation information, including validation characteristics. Additional information on certain methodologies is provided below.
XI. 方法学
II章到第IX章提供了分析方法和分析方法验证资料方面的基本信息,包括验证项目。下文就一些具体的方法给出了说明:
A. High-Pressure Liquid Chromatography (HPLC)
The widespread use of HPLC analytical procedures and the multitude of commercial sources of columns and packings frequently have created problems in assessing comparability. Many of the following points may also apply to other chromatographic analytical procedures.
1. Column
The following characteristics are useful for defining a particular column and, if known, should be included in the analytical procedure description. If method development has indicated that columns from only one commercial source are suitable, this information should be included as part of the analytical procedure. If more than one column is suitable, a listing of columns found to be equivalent should be included.
A.高效液相色谱(HPLC)
HPLC分析方法的广泛应用及色谱柱和柱填充的众多来源都经常会给可比性评估带来很多问题。如下这些要点中,很多都适用于其它色谱分析方法。
1.色谱柱
在定义某一色谱柱时,如下这些性质是很有用的,也应当要包括在分析方法描述中。如果分析方法开发表明只有某一商业来源的色谱柱是适用的,则在分析方法中应当要包括这些资料。如果有多种色谱柱都是适用的话,则应当要包括等效色谱柱列表。
a. Column Parameters
1. Material: glass, stainless steel, plastic
2. Dimensions: length, inner diameter
3. Frit size
4. Filter type
5. Precolumn and/or guard column type, if used
b. Packing Material
1. Particle type: size, shape, pore diameter
2. Surface modification (e.g., bonded surface type, surface coverage, percentcarbon, additional silylation)
3. Recommended pH range for column use
a 色谱柱参数
1.材质:玻璃,不锈钢,塑料
2.尺寸:长度,内径
3.熔封尺寸
4.过滤类型
5.预柱和/或保护柱(如使用)
b.色谱柱填充物
1.颗粒类型:尺寸,形状,孔径
2.表面修饰(如:键合表面类型,表面覆盖,碳比例,甲硅烷基化作用)
3.适用的柱pH范围
2. System Suitability Testing
Each analytical procedure submitted should include an appropriate number of system suitability tests defining the critical characteristics of that system. Criteria for all system suitability testing should be provided. The system suitability tests listed below are defined in CDER=s reviewer guidance on Validation of Chromatographic Methods (November 1994).
1. Tailing factor
2. Relative retention
3. Resolution
4. Relative standard deviation (RSD)
5. Capacity factor
6. Number of theoretical plates
2.系统适应性研究
CDER的评审官指南:色谱方法的验证(1994年11月) 中对如下这些系统适应性实验进行了定义:
1.拖尾因子
2. 相对保留时间
3.分离度
4.相对标准偏差
5.容量因子
6.理论塔板数
The RSD is normally performed at the beginning of the run. However, for assays with lengthy run times or as otherwise justified by the applicant, the reported average may be taken from injections at the beginning and end of the run, or at the beginning, middle, and end of the run.
If an internal standard is used, the minimum acceptable resolution between theinternal standard and one or more active ingredients should be specified. If theanalytical procedure is used to control the level of impurities, the minimum resolution between the active ingredient and the closest eluting impurity, or the two peaks eluting closest to each other, should be given.
一般来说,在仪器开始运行时,进行进样精密度实验,计算其相对标准偏差。然而,对于运行时间很长的分析,或申请者另有理由说明,则可以在运行的开始和结束时,或在运行的开始,中间和结束时进样,然后报告其平均值。
如果用到了内标物,则应当要标明内标物与活性成分(单个或多个)之间的分离度的最低可接受值。如果该分析方法是用于控制杂质水平的,则应当要说明活性成分和最邻近杂质组分,或每两相邻组分的分离度最小可接受值。
3. Operating Parameters
The sequence of injection of blanks, system suitability standards, other standards, and samples should be defined. Flow rates, temperatures, and gradients should be described.Complete details should be provided for the preparation of the mobile phase,including the order of addition of the reagents and the methods of degassing andfiltration. The effect of adjustments in mobile phase composition on retention times should be included in the analytical procedure. The rationale for the use ofprecolumns and/or guard columns should be provided and justified. Any specialrequirements, such as the use of inert tubing or injection valves, should be specified.
3.操作参数
应当要确定空白溶液,系统适应性实验用标准溶液,或其它标准溶液和样品溶液的进样顺序。应当要对流速,温度和梯度洗脱进行描述。
需详细说明流动相的配制,包括试剂的添加顺序,去气和过滤的方法。分析方法中应当要说明流动相组成的调整会对保留时间所产生的影响。如使用了预柱和/或保护住的话,则要进行说明并提供给出合理性解释。如有任何其它要求,如使用了惰性管(inert tubing)或进样阀(injection valve),都应当要对此进行说明。
B. Gas Chromatography (gc)
At a minimum, the following parameters should be included in the description of a gc procedure. Additional parameters should be specified if required by the analytical procedure. If method development has indicated that columns from only one commercial source are suitable, this information should be included as part of the analytical procedure. If more than one column is suitable, a listing of columns found to be equivalent should be included.
一道黑
第17楼2008/11/01
B.气相色谱(GC)
在GC分析方法描述中,至少要包括如下内容。如分析方法中用到了其它的一些参数,对此也应当要进行说明。如果分析方法开发表明只有某一商业来源的色谱柱是适用的,则在分析方法中应当要包括这些资料。如果有多种色谱柱都是适用的话,则应当要包括等效色谱柱列表。
1. Column
---Column dimensions: length, internal diameter, external diameter
---Stationary phase
---Column material (e.g., silica, glass, stainless steel)
--- Column conditioning procedure
2. Operating Parameters
---Gases: purity, flow rate, pressure
---Temperatures: column, injector, detector (including temperature program, if used)
---Injection (e.g., split, splitless, on-column)
---Detector
---Typical retention time and total run time
1.色谱柱
色谱柱尺寸:柱长,内外径
固定相
柱填充料(比如:硅酸,玻璃,不锈钢)
柱调节程序
2.操作参数
载气:纯度,流速,压力
温度:色谱柱,进样器,检测器(包括升温程序,如用到的话)
进样:(比如:分流,不分流,柱头进样(On-column injection)
检测限
典型保留时间和总运行时间。
3. System Suitability Testing
Appropriate system suitability criteria should be defined and included in all analytical procedures.
If an internal standard is used, the minimum acceptable resolution between the internal standard and one or more active ingredient should be specified. If the analytical procedure is used to control the level of impurities, the minimum resolution between the active ingredient and the closest eluting impurity, or the two peaks eluting closest to each other, should be given.
The RSD is normally performed at the beginning of the run. However, for assays with lengthy run times or as otherwise justified by the applicant, the reported average may be taken from injections at the beginning and end of the run, or beginning, middle, and end of the run.
3.系统适应性实验
在所有的分析方法中都应当要有适当的系统适应性要求。
如采用国内标准,则应满足内标和一个或多个组分的最小分离度要求。如果该分析方法是用于控制杂质水平的,则应当要说明活性成分和最邻近杂质组分,或每两相邻组分的分离度最小可接受值。
C. Spectrophotometry, Spectroscopy, Spectrometry and Related Physical Methodologies
These analytical procedures include, but are not limited to, IR spectrophotometry, near IR spectrophotometry (NIR), UV/visible spectrophotometry (UV/Vis), atomic emission and atomic absorption, NMR, Raman spectroscopy, MS, and XRD.
Spectrometric analytical procedures may not be stability-indicating. The bias of the analytical procedure should be evaluated by comparing it with a chromatographic procedure, where appropriate. When manually operated equipment is used, the description of the analytical procedure should include an acceptance criterion for the amount of time that may elapse between sampling and reading. Appropriate system suitability and/or calibration testing is recommended. Validation criteria should include specificity (demonstrating no interference of placebo), linearity, repeatability, intermediate precision, and robustness.
C:分光光度法,光谱法和相关的物理方法
(IR),近红外光说(NIR),紫外可见光谱(UV/Vis),原子发射光谱/原子吸收光谱,核磁共振(NMR),拉曼光谱,质谱(MS)和X-射线衍射(XRD)。
光谱分析方法可能没有稳定性指示能力。必要的话,可能通过与色谱方法的比较来评估光谱分析方法的偏差。如果用到了手动仪器的话,则分析方法描述中应当要有取样(sampling)和读数(reading)之间时间差的可接受标准。建议使用适当的系统适应性实验和/或校准实验。验证标准中应当要包括专属性(说明没有空白干扰),线性,重复性,中间精密度和耐用性。
D. Capillary Electrophoresis (CE)
At a minimum, the parameters listed below should be specified for a capillary electrophoretic analytical procedure. Additional parameters may be included as required by the procedure.
If method development has indicated that capillaries from only one commercial source are suitable, this information should be included as part of the analytical procedure. If more than one capillary is suitable, a listing of capillaries found to be equivalent should be included.
1. Capillary
--Capillary dimensions: length, length to detector, internal diameter, external
diameter
--Capillary material
--Capillary internal coating (if any)
一道黑
第18楼2008/11/01
D:毛细管电泳(CE)
对于一个毛细管电泳分析方法,至少要指明下述参数。如分析方法需要的话,还要包括其它参数。若方法开发研究表明只有某一商业来源的毛细管是适用的话,则在分析方法描述中需包括该信息。如果有多种毛细管柱都是适用的,则要列表说明所有的等效毛细管柱。
1.毛细管柱
毛细管尺寸:长度,至检测器长度,内径,外径
毛细管材质
毛细管内部涂层(如果有的话)
-- Capillary preparation procedure: procedure to be followed before the first use, before the first run of the day, before each run (e.g., flush with 100 millimolar sodium hydroxide, flush with running buffer)
-- Running buffer: composition, including a detailed preparation procedure with
the order of addition of the components
--Injection: mode (e.g., electrokinetic, hydrodynamic), parameters (e.g.,voltage, pressure, time)
-- Detector
--Typical migration time and total run time
--Model of CE equipment used
--Voltage (if constant voltage)
--Current (if constant current)
--Polarity (e.g., polarity of electrode by detector)
2.操作参数
毛细管制备过程:第一次使用前,第一次进样前,每次进样前(比如,用100mmol的氢氧化钠冲洗,或用电流缓冲液冲洗)
电泳缓冲液:组成,包括详细的制备程序及各组分的添加次序。
进样:形式(电动的,水力的),参数(比如,电压,压力,时间等)
检测器
典型的迁移时间和总运行时间
CE仪器型号
电压(如果是恒压的话)
电流(如果是恒流的话)
极性(比如,检测器电极的极性)
Each analytical procedure should include the appropriate system suitability tests defining the critical characteristics of that system. Other parameters may be included at the discretion of the applicant.
3.系统适应性实验
每个分析方法都应当要有适当的系统适应性实验以确定该系统的关键参数。是否包括其它参数,这由申请者来决定。
如用到了内标物,则至少要指明内标物和一个或多个活性组分间分离度的最小可接受值。若该分析方法是用于控制杂质水平的话,则要说明活性组分的最接近流组分间的最小可接受分离度,或每两邻近组分间的最小可接受分离度。
E. Optical Rotation
Optical rotation is used for the measurement of stereochemical purity. Visual polarimeters rely on a monochromatic source, which traditionally was sodium D, but has expanded to virtually any wavelength.
If measurements are to be made at a wavelength other than sodium D, an explanation for selecting the wavelength should be given, along with a comparison of the specific rotation at sodium D and the wavelength to be used. Circular dichroism (CD) spectra may suffice for this purpose. In addition to the provisions of USP <781>, procedures for measurement of specific rotation should include the solvent, concentration, and, for aqueous solutions, the pH to which the solution should be adjusted. The conditions and equipment should be shown to be suitable to confirm the stereochemical identity of a racemate or an enantiomer.
The enantiomeric purity can be expressed as enantiomeric excess (e.e.), using the following formula as an example:
E:旋光度
旋光度用于测定多糖类物质的纯度,通过单色光(钠光)的偏振,但事实上已扩展到使用任一波长的光源。
如不在钠光谱的波长测定的话,则要给出选择该波长的理由,并要对比在钠D和所选用波长下的旋光度。循环二色性图谱能达到这个目的。除了USP<781>中的规定之外,旋光度测定程序还应当要包括溶剂,浓度,及水溶液所要调节到的pH值。所用仪器和条件都要能适用于消旋物或光学异构体的结构确认。
光学异构体纯度可以用异构体过量(e.e.) 来表示,举例如下:
e.e. = 100% * {{M} - [m]}/{[M] + [m]}
where [M] and [m] are the concentrations of the major and minor enantiomers, respectively. This yields values of zero for a racemate and 100 percent for a pure enantiomer. An intermediate concentration gives intermediate values; for example, 97:3 would give an e.e. of 94 percent.
Appropriate system suitability and/or calibration testing is recommended. Validation criteria should include specificity, and intermediate precision.
式中:
和分别是较多光学异构体和较少光学异构体的量。对于消旋物来说,该值为0,若为纯的光学异构体,则为100%。若为中间浓度,则将其表示为中间值;比如,97:3的e.e为94%。
推荐进行适当的系统适应性实验和/或校准实验。验证项目应当包括专属性和中间精密度。
Methodologies Relating to Particle Size Analysis
Particle size analysis is an important element for quality control and regulatory evaluation of certain drug substances and drug products. The normal concepts of validation may differ for particle size methodologies as compared to other analytical methodologies such as HPLC.However, a standard mixture may be used for calibration.
和粒径分析相关的分析方法:
对于有些原料药和制剂的质量控制和官方评审来说,粒径分析是个很重要的因素。
和其它分析方法,比如HPLC相比,粒径分析方法的验证是不尽相同的。然而,在给验证进应用到标准混合物。
Particle size evaluation can include characteristics of size, morphology, surface, and population of particles. The following parameters are useful for describing particle size analysis for characterization of drug substances and drug products.
粒径分析可以包括尺寸特征,形态,表面和粒子群。如下这些参数对于描述界定原料药或制剂所用粒径分析时是很有用的。
1. Particle Size Methods
Types of particle size methods include, but are not limited to:
a. Nonfractionation methods that evaluate an entire population of particles
--Microscopy (optical, electron)
-- Light scattering (dynamic, photon correlation, laser diffraction)
-- Electrozone sensing
-- Photozone sensing
b. Fractionation methods that use physical techniques to separate particles on the basis of size
--Sieving
--Cascade impactor
-- Sedimentation
-- Size exclusion chromatography
一道黑
第19楼2008/11/01
1. 粒径分析方法
粒径分析方法包括,但不局限于:
a. 评估整个粒子群的非分级方法
i. 显微镜检查法
ii. (动态,光子相关,激光衍射)
iii. 电区感应
iv. 辐照感应
b. 根据粒子的尺寸进行分离的物理性分级方法。
i.筛分
ii. 阶式碰撞采样器
iii.沉降法
iv. 尺寸排除色层分析法
2. Calibration and Validation Characteristics
To ensure proper instrument operation, the system should be calibrated according to the manufacturer's and/or the laboratory's specification, as appropriate.
The methods validation usually involves evaluation of intermediate precision and robustness. Assurance should be provided that the data generated are reproducible and control the product's quality. See additional information in sections V and VII.
2.校准和验证
为确保仪器操作的准确性,应根据供应商和/或实验规范对系统进行校准。
分析方法验证经常会包括中间精密度和耐用性(robustness)的评估。应当要确保所得到的数据是可重现的,并能控制产品的质量。更多信息可参见第V和VII章。
G. Dissolution
The equipment used for dissolution is covered by USP <711> or USP <724>. The dissolution procedure description and validation should include the following.
G:溶出度
在USP<711>或USP<724>中说明了溶出度实验所用的设备。溶出度实验分析方法描述及其分析方法验证中应当包括如下资料:
1. Dissolution Medium
A brief discussion of the reasons for selecting the medium.
2. Procedure
A dissolution test consists of a dissolution procedure and method of analysis (automated on-line analysis or manual sampling followed by HPLC analysis). The written procedure should cover the following items:
-- Apparatus
-- Preparation of standard
-- Preparation of sample
-- Method of analysis (e.g., UV, HPLC)
-- Sampling procedure (e.g., intervals, filtration, handling of samples, dilutions)
-- Calculations
-- Acceptance criteria
Regardless of the method of analysis, system suitability criteria should be described.Blank and standard solution spectra or chromatograms should be included.
1.溶解媒介
简要讨论媒介选择的理由。
2.操作程序
(自动在线分析或手动取样然后进行HPLC分析)。书面程序应当要包括如下这几点:
--仪器
--标准品的制备
--样品的制备
--分析方法 (比如,UV,HPLC)
--样品制备(比如,间隔,过滤,样品处理,稀释)
--计算
--合格标准
无论是用什么分析方法,都应当要有系统适应性实验,并要有相应的合格标准。还应当要有空白溶液和标准溶液的光谱或色谱图。
3. Validation Characteristics
Both the dissolution procedure and the method of analysis should be validated.
The time needed for the completion of the sample analysis should be stated in the procedure. Data should be submitted to support the stability of the dissolution sample during the procedure. If filters are used on-line or during sample preparation, appropriate recovery studies should be performed and documented and any bias should be addressed.
3. 验证
无论是溶出程序,还是分析方法都应当要经过验证。
在操作程序中应当要说明完成样品分析所用的时间。还应当要有资料说明样品在实验过程中的稳定性。若进行了在线过滤或在样品制备过程中进行了过滤,则应当要进行适当的回收实验并整理成文件,需说明实验过程中出现的所有偏差。
H. Other Instrumentation
1. Noncommercial Instrumentation
FDA encourages the development and use of the most appropriate instrumentation. However, the use of rare or exotic systems not only places an undue burden on the regulatory laboratory, but also may delay the validation process. When noncommercial instrumentation is used, the instrumentation should be capable of being constructed from commercially available components at a reasonable cost, if possible. For unique methodologies or instrumentation requiring contract fabrication, the applicant's cooperation with the FDA laboratories in helping facilitate duplication of the analytical procedure is important. In addition to design and equipment specifications, complete performance assessment procedures should be provided. Such systems may be found suitable for regulatory use.
H:其它仪器分析方法
1.非商业化仪器
FDA鼓励开发和使用最恰当的仪器。稀有系统的使用不仅给官方实验室带来了过度的负担,也会耽搁验证过程。
若使用非商业化仪器,则应当尽可能地能以经济的商业化配件组装成实验所用仪器。对于需要合同制造的特殊分析方法或仪器,申请者应当要和FDA实验进行合作,以使该分析方法可以重现,这一点是非常重要的。除了需提供仪器设计和仪器规格之年,还应当要提供完整的性能评估程序。这样的系统必须是适用的。
2. Automated Analytical Procedures
The use of automated analytical procedures, although desirable for control testing, may lead to delay in regulatory methods validation because FDA laboratories have to assemble and validate the system before running samples. To avoid this delay, applicants should demonstrate the equivalence of a manual procedure to the automated procedure based on the same principle whenever possible.
2.自动分析方法
自动分析方法的使用会导致官方分析方法验证的延误,即使是非常适用于检测控制的,因为在样品分析之前,FDA实验室必须要对系统进行装配和验证。为了避免这类延误,申请者应当尽可能论证在相同的原理基础上,手动分析方法和该自动分析方法是相当的。
ATTACHMENT A
NDA, ANDA, BLA, AND PLA SUBMISSION CONTENTS
The information relating to analytical procedures and methods validation that should be submitted in NDAs, ANDAs, BLAs, and PLAs is identified below with a cross-reference to the section of this guidance that provides recommendations and/or discussion on the topics.
Information that should be included in the analytical procedures and controls sections
1. Reference standard information Section IV
· Analytical procedures Section III, VI
· Validation data Section VII
· Stress studies Section VII.A.2.c
· Instrument output/raw data for impurities Section VII.A.2.b
· Statistical analysis Section VIII
· Revalidation, as needed Section IX
一道黑
第20楼2008/11/01
附录A
NDA, ANDA, BLA 和PLA申请的内容
NDA,ANDA,BLA和PLA中递交的分析方法和方法验证相关资料如下所示。并标明了本指南中给出了相应建议和/或讨论的章节的章节号。
在分析方法和控制 一章中需包括的资料:
标准品信息,第IV章
分析方法,第III,VI章
分析方法验证资料,第VII章
强降解实验,第VII.A.2.C章
杂质研究的仪器输出/原始资料,第VII.A.2.b章
统计分析,第VIII章
必要的再验证,第IX章
Information that should be included in the methods validation package5
· Contents of the MV Package Section XI
· Representative instrument output/data for stress studies Section VII.A.2.c
·Representative instrument output and raw data for initialand oldest sample of a batch Section VII.A.2.b
分析方法验证中所需包括的资料:
分析方法验证的内容,第XI章
强降解实验的代表性仪器输出/原始资料,第VII.A.2.C章
某一批次最初样品和最老样品的代表性仪器输出/原始资料,第VII.A.2.b章
Information that should be included in the stability section
-- Stress study designs and results Section VII.A.2.b
-- Reference (volume and page number of submission)to instrument output and raw data submitted to the sectiondedicated to analytical procedures and controls Section VII.A 2.c
稳定性章节中所需包括的资料:
强降解实验的实验设计和实验结果,第VII.A.2.b章
参考分析方法和控制章节中所递交的的代表性仪器输出/原始资料,第VII.A.2.C章
ATTACHMENT B
METHODS VALIDATION PROBLEMS AND DELAY
Listed below are examples of common problems that can delay successful validation.
-- Failure to provide a sample of a critical impurity, degradation product, internal standard, or novel reagent
-- Failure to submit well-characterized reference standards for noncompendial drugs
-- Failure to provide sufficient detail or use of unacceptable analytical procedures. For example:
附录B
分析方法验证的问题和延误
下文举列说明了些会延误成功验证的常见问题。
--未能提供关键杂质,降解物,内标物或新试剂的样品。
--未能提供非药典药物的已界定标准品。
--未能提供分析方法的详细描述,或使用和了不可接受的分析方法。比如:
--Use of arbitrary arithmetic corrections
--Failure to provide system suitability tests
--Differing content uniformity and assay analytical procedures without showing
--equivalence factors for defining corrections as required by the current USP chapter
<905> - Uniformity of Dosage Units
-- Failure to submit complete or legible data. For example:
-- Failure to label instrument output to indicate sample identity
-- Failure to label the axes
--随意使用数学校正。
--未能提供系统适应性实验。
--含量均一性分析和含量分析的分析方法是不一样的,且未说明校正用的等效因子,而这是美国药典第<905>章:剂型的均一性 所必需的。
--未能提供完整清晰的数据。比如:
--未能标注仪器输出。
--未能标注坐标。
! Inappropriate shipping procedures. For example:
-- Failure to properly label samples
-- Failure to package samples in accordance with product storage conditions
-- Inadequate shipping forms (e.g., missing customs form for samples from outside the United States)
! Failure to describe proper storage conditions on shipping containers
不合理的运送方式。比如:
-- 未能正确标注样品
-- 未能根据产品储存条件来包装产品。
--货运不完备(比如,没有国外样品的报关单)
--未在运送包装上标明适当的储存条件。