fslcx
第25楼2010/04/07
Tune report for VARIAN 1200L
Time: 22:59:33 Tune file: FC-43 EI (Tuned on 1 Apr 2010)
Ionization mode:EI Scan Optimization: Standard gc (Fixed detector)
Exact Measured Peak Relative Peak Valley Resolution
mass mass height height width
(amu) (amu) (mV) (%) (amu @ 50%) (% of iso) (m/delta-m)
18.0 18.1 2.140 100.000 0.68 100 27
28.0 27.9 0.793 37.053 0.74 96 38
69.0 69.0 1.117 52.196 0.71 100 97
219.0 219.0 1.890 88.342 0.71 99 311
502.0 502.1 0.110 5.149 0.97 12 520
(Detector = 1245 V)
Tuning of quad 1 in positive mode completed.
Time: 23:05:27 Tune file: FC-43 EI (Tuned on 1 Apr 2010)
18.0 18.0 2.245 66.556 0.67 100 27
28.0 27.9 0.738 21.879 0.71 80 39
69.0 69.0 3.372 100.000 0.66 100 105
219.0 219.0 1.452 43.052 0.73 69 301
502.0 501.9 0.090 2.673 0.97 13 519
(Detector = 1245 V)
Tuning of quad 3 in positive mode completed.
28 absolute size (mV - EDR on)
- Limit 20.0 - Measured 0.0
- OK!
28/32 Ratio
- Air 3.5:1 - Measured 3.5:1
- Indicative of Air
28/18 Ratio
- Limit 1.2:1 - Measured 1.0 :1
- OK!
- No significant leak detected -
- OK -
调谐结果报告正常,但是几个质量数的响应均掉了两个数量级,原因未知。