    第12楼2011/02/25

    For your 2nd problem, please also check your follow cell (air bubble, stain, etc.). If you have air bubble (long time no use) in the cell, light intensity from the D2 lamp will be much lower, give you the error message.

0
    +关注 私聊
  • jy010422

    第13楼2011/02/26

    But how to eliminati the bubble?Thanks for your guidance!

    zlignc(zlignc) 发表:For your 2nd problem, please also check your follow cell (air bubble, stain, etc.). If you have air bubble (long time no use) in the cell, light intensity from the D2 lamp will be much lower, give you the error message.

0
    +关注 私聊
  • zlignc

    第14楼2011/02/26

    1. Create some back pressure after the dectector and suddenly release the pressure. Repeat this several time.
    2. Flow pure organic solvent with how flow rate through your flow cell.
    3. Use syringe to push methanol through the the flow cell.

    You can try the above methods. You can also take out the flow cell to check the presend of bubble.

0
    +关注 私聊
  • dony1985

    第15楼2011/02/26

    还搞英文啊!牛B!
    兰州哪个地区的,还是800报修吧。

0
    +关注 私聊
  • hfang2005

    第16楼2011/02/27

    咋整起英文了呢? 真牛X啊!
    两个问题整一块去了.
    1. 实际有的硬件模块和软件中的配置不一样. 重新做软件系统配置(system configuration),或检查有哪个模块电源望了打开.
    2. 氘灯坏了,换灯!

0