    药典的要求看来确实不高,但这应该就是一个最低要求吧?想请教根据实际使用经验,信噪比控制在什么样会较好呢?

    picard(yuan314159) 发表:比较的话,可以按药典要求,当然药典的要求是相当宽松的。

    Table 1.Recommended Near-IR Instrument Specificationsa

    Wavelength UncertaintySRM1920a peaksboccur at 1261,1681,and 1935nm
    Tolerances±1nm at 1200nmor ±8cm-1at 8300cm-1
    ±1nm at 1600nmor ±4cm-1at 6250cm-1
    ±1.5nm at 2000nmor ±4cm-1at 5000cm-1
    Photometric LinearityAOBSvs AREFat 1200,1600,and 2000nm;c
    slope =1.0±0.05;intercept =0.0±0.05
    Spectrophotometric Noisemeasured for 100-nm (300cm-1)segments between 1200and 2200nm (8300and 4500cm-1)
    Average RMSfor measurements
    at high-light flux
    less than 0.3×10-3;no RMSnoise greater than 0.8×10-3
    Average RMSfor measurements
    at low-light flux
    less than 1×10-3;no RMSnoise greater than 2.0×10-3

    a Amaximum nominal instrument bandwidth of 10nm at 2500nm or 16cm-1at 4000cm-1is appropriate for most applications.

    b The nominal 1935-nm peak is sensitive to instrument bandwidth.Use the wavelength value supplied with SRM1920a at the appropriate instrument bandwidth to determine wavelength uncertainty.

    c AOBSis the observed absorbance,and AREFis the tabulated absorbance of the reference reflectors at each of the three specified wavelengths.