乳品中三聚氰胺检测方案(液质联用仪)
检测样品
液体乳
检测项目
非法添加
参考标准
DB32/T 1163 鸡肝中金刚烷胺残留量的测定 液相色谱-串联质谱法
关联设备
共1种
下载方案
The high resolution, accurate mass Exactive mass spectrometer was shown to be sensitive in detecting <0.1 ppb melamine (0.5 pg on column) in neat standard, and response is linear from 0.1 to 100 ppb. The error for mass accuracy is <2 ppm with lock mass.
Milk samples prepared by dilute-and-shoot showed severe ion suppression that was reduced with a simple 1-minute isocratic HILIC LC separation, after which a consistent response factor of 0.3-0.5 for each sample matrix was obtained for quantitation.
Quantitation limits were less than 44, 65, and 110 ppb for coffee cream, infant formula and 3-in-1 instant coffee mix, respectively, exceeding the requirement of 250 ppb LOQ as stated by US FDA for infant formula.
赛默飞质谱分析为您提供《乳品中三聚氰胺检测方案(液质联用仪)》,该方案主要用于液体乳中非法添加检测,参考标准《DB32/T 1163 鸡肝中金刚烷胺残留量的测定 液相色谱-串联质谱法》,《乳品中三聚氰胺检测方案(液质联用仪)》用到的仪器有赛默飞Q Exactive Plus LCMS 系统。
我要纠错