    Potential Metal Interferences on the
    Determination of Arsenic(III)

    The interference of 16 metal species on
    the determination of As(III) was tested by
    spiking synthetic samples containing 10 μg/L
    As(III) and 10 μg/L As(V) with 0.0001 to
    1,000 mg/L Al, Cd, Co, Cr(III), Cr(VI),
    Cu(II), Fe(III), Fe(II), Mn, Ni, Pb, Sb(III),
    Sb(V), Se(IV), Se(VI), and Zn and
    determining the As(III) concentration using
    the HGAAS procedure (without cationexchange
    treatment). Arsenic(III) recoveries
    for samples spiked with Al, Co, Cr(III),
    Fe(II), Mn, Ni, and Zn ranged from 84 to 107
    percent over the entire concentration range
    tested (fig. 5).
    Low As(III) recoveries occurred when
    Cd exceeded 25 mg/L, Cr(VI) exceeded 0.03
    mg/L, Cu(II) exceeded 2.0 mg/L, or Fe(III)
    exceeded 1.0 mg/L. Cadmium and Cu(II) are
    likely reduced by NaBH4 in the FIAS reaction
    chamber and, consequently, arsine production
    is limited by the available NaBH4.
    Chromium(VI) and Fe(III) inhibit arsine
    production by either competing with As(III)
    for the available NaBH4 or oxidizing As(III)
    to As(V). Arsenic(III) is quickly oxidized as
    Fe(III) undergoes photoreduction (Emett and
    Khoe, 2001). These synthetic samples were
    prepared in clear volumetric flasks and
    exposed to light. Therefore, the low As(III)
    recovery in the presence of only Fe(III) may
    be due to photochemical reactions with
    Fe(III) prior to the As(III) determination.
    However, when Fe(II) concentrations were
    twice the concentration of Fe(III), low As(III)
    recoveries occurred only when Fe(III)
    exceeded 10 mg/L (fig. 6). Iron(II) relieves
    the interference from Fe(III), within a limited
    Fe(III) concentration range, by reacting with
    free radicals generated as Fe(III) undergoes
    photoreduction (Emett and Khoe, 2001).
    Among the hydride-forming metals, Sb(III) at
    concentrations greater than 0.1 mg/L, Sb(V)
    greater than 0.4 mg/L, and Se(IV) greater than
    0.02 mg/L interfered with As(III)
    determinations, whereas Pb and Se(VI) did
    not (fig. 7). Antimony(III), Sb(V), and Se(IV)
    suppress AsH3(g) production by reacting with
    the available NaBH4. An additive effect of
    interfering metals on the determination of
    As(III) was demonstrated by adding
    increasing amounts of Fe(III) to a solution
    containing 0.5 mg/L Cu and 10 μg/L As(III)
    (fig. 8). Low As(III) recoveries occurred
    when the molar ratios of metals to As(III)
    were: Cu(II) greater than 120, Fe(III) greater
    than 70, Cr(VI) greater than 2, Cd greater
    than 800, Sb(III) greater than 3, Sb(V) greater
    than 12, or Se(IV) greater than 1. The sample
    could not be diluted to an As(III)
    concentration (As(III) less than 20 μg/L)
    below which these interferences were absent.
    Copper(II) and Fe(III) are of primary concern
    because water generated from acid mine
    drainage potentially contains high
    concentrations of Cu(II) and Fe(III).

0
0
0
0
0
    +关注 私聊
  • 第20楼2005/04/12

    Cation Exchange Separation of
    Iron(III) and Copper

    Interfering metals can be removed using
    cation-exchange resin prior to the
    determination of As(III). The relative
    selectivity for AG 50W-X8 cation-exchange
    resin is 2.95 for Cd, 2.9 for Cu(II), and 1.0 for
    H+ (Bio-Rad, 2003). Therefore, as long as the
    capacity of the resin (1.7 millieqivalents
    (meq) per mL of resin) is not exceeded, Cd
    and Cu(II) are expected to be removed from
    solution. The relative selectivity for Fe(III) is
    not published in the AG 50W instruction
    manual.
    Laboratory experiments done for this
    report demonstrated that AG 50W-X8 cationexchange
    resin in the H+ form removed large
    amounts of Cu(II) and Fe(III) while
    maintaining the existing As(III)/As(T) ratio
    (table 1). When the As(III) concentration is
    high relative to that of the interfering metal,
    the sample may be diluted before analysis to a
    Cd, Cu(II), Cr(VI), Fe(III), Sb(III), or Se(IV)
    concentration that does not interfere with the
    As(III) determination. Chromium(VI) and the
    Se and Sb oxyanions will not be removed by
    cation exchange.