    Accuracy and Time Stability of
    Arsenic Redox Species in Acid Mine
    Waters


    Accuracy of Arsenic(T) Determinations

    Acid mine water may contain up to 31
    major metal species with concentrations that
    can vary over several orders of magnitude
    (Nordstrom and Alpers, 1999). Consequently,
    interferences are likely during As redox
    determinations by HGAAS for acid mine
    waters. Samples of water collected from the
    Reynolds Adit and a monitoring well near the
    Chandler Adit on the Summitville Mine site,
    the Richmond Mine, and the Penn Mine were
    analyzed for many dissolved constituents
    including As(T) and As(III). Concentrations
    of potential interfering metals for these
    samples are listed in Table 2. The samples
    contain elevated concentrations of metals, as
    determined using inductively coupled plasmaoptical
    emission spectroscopy (ICP-OES) and
    colorimetric determinations for Fe redox
    species (To and others, 1999).
    Accuracy of the As(T) method was
    evaluated by performing a standard addition
    on samples collected from the Summitville
    Mine, Colo. (Reynolds Adit and a monitoring
    well near Chandler Adit); the Richmond
    Mine, Calif.; and the Penn Mine, Calif. Spike
    recoveries for the samples ranged from 96 to
    101 percent (fig. 9). Twenty samples from the
    Summitville Mine site, 6 from the Richmond
    Mine, and 3 from the Penn Mine were
    analyzed by ICP-OES for As(T).
    Concentrations compared well with As(T)
    concentrations determined by hydride
    generation (fig. 10). All As(T) determinations
    were performed without cation-exchange
    separation. As another measurement of
    accuracy for As(T) determinations, USGS
    standard reference water samples (SRWS)
    AMW4, T143, T155, and T159 (Farrar, 2000)
    were analyzed as unknowns along with the
    acid mine water samples. Measured
    concentrations compare well to the most
    probable values (MPV) (table 3).

0
0
0
0
0
0
    +关注 私聊
  • 第27楼2005/04/12

    Accuracy of Arsenic(III) Determinations

    Accuracy of As(III) determinations, both
    with and without cation exchange, was
    estimated by spiking samples collected from
    the Summitville Mine (Reynolds Adit and
    Chandler Adit), the Richmond Mine, and the
    Penn Mine with As(III). Iron(III) and Cu(II)
    interfere with the determination of As(III) in
    samples collected from Summitville, but not
    with the Richmond or Penn Mine samples (fig
    11). Spike recoveries for Summitville Mine
    samples not separated by cation exchange
    were 44 to 53 percent, whereas recoveries for
    the same samples separated by cation
    exchange were 102 to 103 percent. The
    Richmond and Penn Mine samples were not
    separated with cation exchange resin because
    the ratios of Fe(III) and Cu(II) to As(III) were
    low enough that samples could be diluted to
    Fe(III) and Cu(II) concentrations that did not
    interfere with the determination. The As(III)
    concentrations were underestimated in the
    Summitville Mine samples not separated by
    cation exchange (fig. 12). No USGS standard
    reference water sample exists for As redox
    species.

0
0
0
    +关注 私聊
  • 第30楼2005/04/12

    Time Stability of Arsenic Redox Species
    The time stability of As(III) was
    monitored in 45 surface and ground water
    samples from Yellowstone National Park,
    Wyo.; Questa Mine site, N. Mex.,
    Summitville Mine site, Colo., Richmond
    Mine, Calif., Penn Mine, Calif., Ester Dome,
    Alaska, Fallon, Nev., Mojave Desert, Calif.,
    and Kamchatka, Russia. Samples were
    filtered through a 0.1-μm membrane, acidified
    to pH <2 with HCl, and stored in opaque
    bottles at 4°C. Samples were reanalyzed as
    many as 15 months after the initial
    determination. The samples containing Cu(II)
    or Fe(III) concentrations that interfered with
    the As(III) determination were separated
    using cation exchange. The change in As(III)
    is represented by plotting percent difference
    in As(III) concentration as a function of the
    initial As(III)/As(T) ratio. The equation used
    to calculate percent difference (Δ%) is:

    The As(T) concentration ranged from 0.006 to
    33 mg/L and the initial As(III)/As(T) ratio
    ranged from 0.01 to 1.0 (fig. 13). The curved
    line is a Gaussian Fit. The average percent
    difference (Δ%) for all samples was 0.2 with a
    standard deviation of 7.