This is my background. According to the certificate, 80%ACN / 20%H2O
The second picture is my blank (MeOH and H2O). How can I find what compound it is by mass? I am new to masshunter, don't know how to search the library. I can only use chemistation on gcMS to search the library.
胜邪
第6楼2014/05/31
LCMS has not library that like NIST gcms used.
DAD signal is cleanly.
149.1m/z,esi scan signal at 9.647min. its mean the line pipe in system are dirty.
you can try replace the pipe and column, then purge the system repeatedly.