In ADVANTAGE software, there is an option to set the Ion gun sputtering as AUTO meaning it will set the Ion gun sputtering area to be 5X the X-ray spot size.
Normally if no change of X-ray spot say 400um, then it means sputtering is 2mmx2mm.
I mentioned about this because if you change X-ray spot size > then sputtering area change > then sputtering rate change.
Next to know the sputtering rate, usually the most common way is to do a sputter depth profile experiment on a STANDARD sample (e.g. 100nm SiO2 on Si, or Ta2O5...etc). I think instrument vendors should do it for users on such reference sputtering rate checking when performing maintenance calibration on the instrument?
摆度
第3楼2014/10/25
Yes, the instrument user told you the truth. This sputtering rate is based on Ta2O5 not your sample.
Sputtering rate for each different material will be surely different and only way to know the TRUE sputtering rate on your sample is as if you have a standard known thickness of same material > then you try it out to sputter it through to determine the TRUE sputtering rate.
But as you know, it is almost impossible to have ALL standard with known thickness for ALL kinds of sample in real-life. So usually a common way to do is reference to some STANDARD material sputtering rate and at least it can serve as some "reference". So take your example, you could still use the sputtering rate of 0.51nm/min in the software menu you capture > and then calculate the depth. But if you are writing some paper or something you need to mention that you are referencing to a sputtering rate on Ta2O5.