Hansen
第1楼2006/04/17
Drive amplitude excede the limit mean that detector can not get the signal from cantilever reflection. You can check
Please make sure the red laser spot is on cantilever from CCD monitor and sum value is ok.
If Not. It mean
(a) It is the problem of plug or wire between head and stage. You can check if it is open or short by using electric meter. Please check manual the meaning of the pin.
(b) Piezo on tip holder is broken. This can be checked if there is sum value during laser alignment.
(C) Sometimes it is software bug. please install software again.
If yes.
(d) You just get sum and RMS value ok. But laser is on chip set , not cantilever end. This is becasue you use different kind of tip or whole cantilever is not on the right position.This happened very much.
(e) Sometimes it is software bug. please install software again.
(f) The controller circuit has problem. But this does not happened very often.
Please check what the situation before and during cantilever tunning.
Thanks
Hansen
第3楼2006/04/19
From your explanation. Please check follows.
(e) Sometimes it is software bug. please install software again.
(f) The controller circuit has problem. But this does not happened very often.
You should have bcakup CD to install software again. If it can not work, to find Veeco China for supporting.
Thanks
tmou
第6楼2006/06/20
看一下你的Drive Amplitudes是多少?我估计已经是几伏特了。手工减小到100mV,再试试。
或者,不要使用AutoTune,改为手动搜索