  • iceriver_99
    2006/04/17
  • 私聊

扫描探针显微镜SPM/AFM

  • 我们用的是DI3100,现在contact模式下能够正常使用,但是在Tapping模式下,调节悬臂时总是出现错误提示“Drive amplitude exceeded safe limits while trying to attain target amplitude. This could be due to a broken or misaligned cantilever”。我换过几个悬臂,清理了放置悬臂的凹槽,调节好了激光束,但是还是不行,总是出现这个提示,这样的话Tapping模式就没办法用了,希望有经验的朋友帮帮忙,看看到底是什么原因!在此先谢谢各位!!!
    +关注 私聊
  • Hansen

    第1楼2006/04/17

    Drive amplitude excede the limit mean that detector can not get the signal from cantilever reflection. You can check

    Please make sure the red laser spot is on cantilever from CCD monitor and sum value is ok.

    If Not. It mean

    (a) It is the problem of plug or wire between head and stage. You can check if it is open or short by using electric meter. Please check manual the meaning of the pin.

    (b) Piezo on tip holder is broken. This can be checked if there is sum value during laser alignment.

    (C) Sometimes it is software bug. please install software again.

    If yes.

    (d) You just get sum and RMS value ok. But laser is on chip set , not cantilever end. This is becasue you use different kind of tip or whole cantilever is not on the right position.This happened very much.

    (e) Sometimes it is software bug. please install software again.

    (f) The controller circuit has problem. But this does not happened very often.

    Please check what the situation before and during cantilever tunning.

    Thanks

0
    +关注 私聊
  • iceriver_99

    第2楼2006/04/18

    谢谢这位朋友!现在的情况是:
    1.激光束能射到针尖上,sum值也在正常范围内
    2.contact模式下能正常使用
    3.换过几个悬臂,还是不行
    4.在contact模式下做过电场的原位观察,即在样品两端加电压,但是样品不是导电材料,无电流,最高加到200V,不影响contact模式下的成像。

0
    +关注 私聊
  • Hansen

    第3楼2006/04/19

    From your explanation. Please check follows.

    (e) Sometimes it is software bug. please install software again.

    (f) The controller circuit has problem. But this does not happened very often.

    You should have bcakup CD to install software again. If it can not work, to find Veeco China for supporting.

    Thanks

0
    +关注 私聊
  • hitttr

    第4楼2006/05/24

    试着减小振源幅度,看看有没有效果。

    iceriver_99 发表:谢谢这位朋友!现在的情况是:
    1.激光束能射到针尖上,sum值也在正常范围内
    2.contact模式下能正常使用
    3.换过几个悬臂,还是不行
    4.在contact模式下做过电场的原位观察,即在样品两端加电压,但是样品不是导电材料,无电流,最高加到200V,不影响contact模式下的成像。

0
0
    +关注 私聊
  • tmou

    第6楼2006/06/20

    看一下你的Drive Amplitudes是多少?我估计已经是几伏特了。手工减小到100mV,再试试。
    或者,不要使用AutoTune,改为手动搜索

    iceriver_99 发表:我们用的是DI3100,现在contact模式下能够正常使用,但是在Tapping模式下,调节悬臂时总是出现错误提示“Drive amplitude exceeded safe limits while trying to attain target amplitude. This could be due to a broken or misaligned cantilever”。我换过几个悬臂,清理了放置悬臂的凹槽,调节好了激光束,但是还是不行,总是出现这个提示,这样的话Tapping模式就没办法用了,希望有经验的朋友帮帮忙,看看到底是什么原因!在此先谢谢各位!!!

0
0