离子色谱忘记关机

离子色谱(IC)

  • 戴安的1500外接rfc30,周五样多一直开着,后来忘了关机。周一去实验室发现主机和rfc都已关闭只是还通着电。后来正常开机发现电导很高15000多,检查了其他部分,没有发现异常,接着以正常流速压力浓度运行,经过连续2天20多个小时的运行,绝对电导降到了1.8左右,今天继续运行。

    总结这次的问题:

    程序文件:Pressure.LowerLimit = 200 [psi]
    Pressure.UpperLimit = 3000 [psi]
    %A.Equate = "20mM KOH"
    Pump_InjectValve.State LoadPosition
    Pump_ECD.Data_Collection_Rate = 5.0 [Hz]
    CellTemperature.Nominal = 35.0 [°C]
    ColumnTemperature.Nominal = 30.0 [°C]
    Suppressor_Type = ASRS_4mm
    ; Pump_ECD.Carbonate = 0.0
    ; Pump_ECD.Bicarbonate = 0.0
    ; Pump_ECD.Hydroxide = 20.0
    ; Pump_ECD.Tetraborate = 0.0
    ; Pump_ECD.Other eluent = 0.0
    ; Pump_ECD.Recommended Current = 50
    Suppressor_Current = 50 [mA]
    Channel_Pressure.Step = 0.20
    Channel_Pressure.Average = Off
    Flow = 1.00 [ml/min]

    -2.300 Pump_ECD_Relay_1.Closed Duration=138.00

    0.000 Pump_ECD.Autozero
    ECD_1.AcqOn
    Channel_Pressure.AcqOn
    Pump_InjectValve.InjectPosition Duration=30.00

    11.40 ECD_1.AcqOff
    Channel_Pressure.AcqOff

    End
    后来发现高纯水在130ml的时候主机已经关闭。想来是压力超下限了。而rfc30后来经咨询厂家工程师,模式在ttl control下自动关机对rfc30也没有影响。

    所以这次的意外似乎影响不大。
    +关注 私聊
  • lczhfj

    第1楼2012/07/18

    有时意外情况出现让人手足无措,但未必都会造成大损伤,因有些仪器有保护功能。但细心仍是需要的,尽力避免不规范操作。此可供大家参考。

0
    +关注 私聊
  • seafling

    第2楼2012/07/22

    应该没什么影响,流动相流感的时候,管子里是空气,淋洗液发生器即使开着也不会起作用,仪器无法构成电解回路

0