我用的是岛津2010,今天重新调谐后,出现这个错误信息,请问大家这是什么原因造成的呢? 0923)Mass Calibration was performed up to m/z 502, because the intensity of m/z 614 was low. Cause: The intensity of m/z 614 was lower than the intensity required for the Mass Calibration, so it was performed by using data up to m/z 502.
Remedy: There is no necessity for a measure if the needed mass range is lower than m/z 500. However, check the Mass Calibration Result of m/z 614, and if necessary it is recommended to perform the maintenance for the ion source, because m/z 614 was not used for the Mass Calibration. The situation may be improved by removing the contamination of the ion source. When the needed mass range is upper than m/z 500, it is also recommend to perform the maintenance for the ion source if necessary.