szdavid1
第2楼2006/06/19
多谢回复,我们CE的方法就是下面的方法, 长度是1米,您看看,多谢!
IDENTIFICATION
A. It gives the appropriate response when examined using
the conditions described under Assay.
B. Capillary zone electrophoresis (2.2.47)
Test solution. Dilute the substance to be examined
with water R to obtain a concentration of 1 mg/ml.
Desalt 0.25 ml of the solution by passage through a
micro-concentrator cartridge provided with a membrane
with a molecular mass cut-off of not more than 10 000.
Add 0.2 ml of water R to the sample and desalt again.
Repeat the desalting procedure once more. Dilute the
sample with water R, determine its protein concentration
as described under Tests and adjust to a concentration of
approximately 1 mg/ml with water R.
Reference solution. Dissolve the contents of a vial of
erythropoietin BRP in 0.25 ml of water R. Proceed with
desalting as described for the test solution.
Capillary:
— material: uncoated fused silica,
— size: effective length = about 100 cm, Ø = 50 μm,
Temperature: 35 °C.
CZE buffer concentrate (0.1 M sodium chloride, 0.1 M
tricine, 0.1 M sodium acetate). Dissolve 0.584 g of
sodium chloride R, 1.792 g of tricine R and 0.820 g of
anhydrous sodium acetate R in water R and dilute to
100.0 ml with the same solvent.
1 M putrescine solution. Dissolve 0.882 g of putrescine R
in 10 ml of water R. Distribute in 0.5 ml aliquots.
CZE buffer (0.01 M tricine, 0.01 M sodium chloride,
0.01 M sodium acetate, 7 M urea, 2.5 mM putrescine).
Dissolve 21.0 g of urea R in 25 ml of water R by warming
in a water-bath at 30 °C. Add 5.0 ml of CZE buffer
concentrate and 125 μl of 1 M putrescine solution. Dilute
to 50.0 ml with water R. Using dilute acetic acid R,
adjust to pH 5.55 at 30 °C and filter through a 0.45 μm
membrane filter.
Detection: spectrophotometer at 214 nm.
Set the autosampler to store the samples at 4 °C during
analysis.
Preconditioning of the capillary: rinse the capillary for
60 min with 0.1 M sodium hydroxide filtered through a
0.45 μm membrane filter and for 60 min with CZE buffer.
Apply voltage for 12 h (20 kV).
Between-run rinsing : rinse the capillary for 10 min with
water R, for 5 min with 0.1 M sodium hydroxide filtered
through a 0.45 μm membrane filter and for 10 min with
CZE buffer.
Injection: under pressure or vacuum.
Migration: apply a field strength of 143 V/cm(15.4 kV
for capillaries of 107 cm total length) for 80 min, using
CZE buffer as the electrolyte in both buffer reservoirs.