准确测定系列8:Environmental Contamination

ICP光谱

  • 8. Environmental Contamination

    During the course of a sample preparation, contamination can occur from the environment, the reagents, the apparatus, and even the analyst. In many cases the blank determines the lower limit of detection. The next several parts of this guide deal with contamination.


    Overview
    Environmental contamination is caused by particulate and/or gaseous matter in the air. It has been reported that air in an analytical laboratory can contain up to 200 µg/m3 of particulate matter containing Ca, Si, Fe, Na, Mg, K, Tl, Cu, Mn, and lesser amounts of other elements1. In addition, this reference states that normal rural area airborne particle counts have been reported to be 1,400,000/m3 for particles greater than 0.5 µm and that normal metropolitan area particle counts have been reported to be 53,000,000/m3 for particles greater than 0.5 µm.


    Reducing Environmental Contamination:
    Use the chemical blank as the performance criteria. Dealing with environmental contamination is expensive. Therefore, reduce the blank to the lowest possible value by eliminating contamination from the sample container, the chemist, and sample preparation apparatus. Achieve this by doing the following:

    Separate the sample preparation and handling area from all other areas.
    Eliminate as many foreign objects as possible from the preparation area (i.e. - Motorized objects such as stirring plates).
    Use all-plastic hoods.
    Keep the sample out of the hood area as much as possible.
    The effect of the laboratory atmosphere on lead blank levels is shown in Table 8.1.

    Table 8.1: Effect of Laboratory Atmosphere on Pb Blank Levels

    Hoods, which enhance atmospheric exposure and hence contamination, are necessary for the safety of the analyst. Atmospheric contamination has been reduced significantly through the use of "clean rooms" that use HEPA filtered air. HEPA (High Efficiency Particulate Air) filters are 99.99 % efficient in removing particulates down to 0.3 µm.

  • 该帖子已被管理者-设置为精华,下面是奖励记录:
    +关注 私聊
  • 第1楼2005/03/30

    A Positive Pressure Filtered Air "Clean" Room is shown in the following architectural diagrams
    Diagram 8.1: Clean Room Layout (Top View)


    Diagram 8.2: Clean Room Layout (Front View)



    Diagram 8.3: Clean Room Layout (Side View)

0
    +关注 私聊
  • 第2楼2005/03/30

    The effectiveness of clean rooms in eliminating environmental contamination is illustrated in Tables 8.2 and 8.3 which show a significant reduction in the blank.

    Table 8.2: Analytical Results for Trace Impurities in Ultra-Pure Nitric Acids2
    (Concentrated in Conventional Chemical and Clean Rooms)


    Table 8.3: Analytical Results for Trace Impurities in Ultra-Pure Water Samples2
    (Concentrated in Conventional Chemical and Clean Rooms)



    For budgets that cannot handle the cost associated with clean room facilities there have been several ingenious designs that are effective in dealing with environmental contamination. Diagram 4.4 illustrates some of these designs using equipment that is relatively inexpensive and easily obtained.

0
    +关注 私聊
  • 第3楼2005/03/30

    Diagram 8.4: Evaporation Chamber Designs1


    Avoiding Environmental Contamination:
    Teflon coat or epoxy paint the (hopefully few) metal objects in the laboratory. Epoxy paint the walls. Walls should be washed down weekly and repainted annually.
    Use special adhesive floor pads for all openings to the laboratory.
    Use disposable plastic gloves and disposable paper laboratory coats.
    Vinyl covered floors should be mopped daily. Floors with drains allow for easier daily washing.

0
  • 该帖子已被管理者-设置为精华,下面是奖励记录:加5积分,加5声望
    +关注 私聊
  • 第4楼2005/03/31

    环境沾污,注意痕量分析时空气、房间的装修都有可能的

0
    +关注 私聊
  • 第5楼2005/04/05

    所以现在不少分析实验室,在接受国外机构认定时,需要建设为超净实验室。