ICP错误提示,你遇到过哪些?

  • chun870605
    2013/08/24
  • 私聊

ICP光谱

  • PE的ICP错误提示


    切割气压力不够

    Plasma shear gas is not flowing.

    A shear gas (compressed air or nitrogen) must be flowing

    while operating the plasma to prevent damage to the system.

    Check the following:

    ? Make sure the shear is connected and turned on.

    ? Make sure the pressure is set to a minimum of 415 kPa

    (60 psi).

    After performing these checks try again to operate the

    plasma. If you are not successful after several trials, contact

    PerkinElmer service and arrange a service visit.

    [0514]

    ----------------------------------------------------------------------------

    氩气压力不够



    ---------------------------------------------------------------------------------------------------------------------------------

    这个经常遇到,玩具测试应该都没有氮气或氩气吹扫光室吧



    ------------------------------------------------------------------------------------------------------------------------------

    进样系统漏气,废液拍不出去,等离子不稳定

    Plasma Impedance (Plasma-Z) exceeds working range.

    Check that all sample introduction tubing is properly

    connected.

    Check/perform maintenance on the sample introduction

    system including the torch, injector, injector adapter, and

    nebulizer.

    Check/replace all torch O-Rings if necessary.

    Change the method to reduce sample or vapor loading to

    improve analytical performance.

    [0573]

    ------------

    Plasma went out unexpectedly.

    If this happened during ignition, try the following:

    ? Observe the plasma discharge during ignition through the

    viewing window. If the plasma lights, but then extinguishes

    right away, check the argon flows, torch condition, and test

    for air leaks.

    ? If the sample is very volatile, try a less volatile sample or

    switch to water.

    ? If analyzing volatile samples, a higher power setting and

    lower nebulizer gas flow rate may be required.

    If this happened during normal operation then the plasma

    detection circuit that monitors the plasma is the most likely

    problem. If you continue to have problems, contact

    PerkinElmer service and arrange a service visit.

    [0516]

    ----------------------------------------------------------

    切割气电磁阀出问题




    ------------------------------------------------------------------------------------------------------------------------------------

    冷水机问题,纯净水不够,压力和温度不稳定




    -----------------------------------------------------------------------

    仪器突然断电之后出现的问题,第一次遇到,重启主机电源解决了







    ---------------------------------------------------------------------------------------------------------------------------

    印象中这两个问题遇过两三遍,重启主机电源解决

    The spectrometer has a serious hardware problem.

    A service visit may be required. Consult the Spectrometer

    page of the Diagnostics window for additional information

    about the problem.

    You may be able to correct the problem by reinitializing the

    spectrometer. Click on the Reconnect button on the

    Spectrometer page of the Diagnostics window. The amount

    of time required will depend on the spectrometer being

    initialized.

    [0153]

    -----------



    --------------------------------------------------------------------------------------------------------------------------------------------------------------

    这问题比较奇怪,进行了很多操作都不行,最后换了一根全新的矩管解决了,之前的那两根矩管裂缝太大不能使用

    Current through power transistor out of tolerance.

    This problem may be due to a damaged transistor requiring

    a service visit or it may be due to improper plasma

    conditions.

    Try the following:

    ? Observe the plasma discharge during ignition through the

    viewing window. If the plasma lights, but then extinguishes

    right away, check the argon flows, torch condition, and test

    for air leaks.

    ? If the plasma is being blown out of the torch or is unstable

    during ignition, then check for proper argon flow and air

    leaks.

    ? If the plasma is stable during ignition, but still extinguishes,

    one of the transistors may have been damaged and must be

    replaced by service.

    Try again to ignite the plasma. If you are not successful after

    several trials, contact PerkinElmer service and arrange a

    service visit.

    [0506]

    还有很多遇到过的问题当时没记下来,以后慢慢更新
  • 该帖子已被版主-tang566加5积分,加2经验;加分理由:归类总结
    +关注 私聊
  • 千层峰

    第1楼2013/08/25

    建议整理一下,参加原创。
    仪器还是要好好保养护理。出问题的耗材该换就换。

0
    +关注 私聊
  • tang566

    第2楼2013/08/26

    应助达人

    安捷伦的错误提示,RF冷却风量不足。



    一般是由于风压感应器的测风口有微堵的情况。

0
    +关注 私聊
  • 依风1986

    第3楼2013/08/26

    应助达人

    这个还没碰到过,TANG566版主不是找安捷伦维修过吗?维修细节怎么样?我知道安捷伦ICP点火要两次,一般第一次点火后熄灭会提示F1245

    tang566(tang566) 发表:安捷伦的错误提示,RF冷却风量不足。



    一般是由于风压感应器的测风口有微堵的情况。

0
    +关注 私聊
  • tang566

    第4楼2013/08/26

    应助达人

    维修细节有拍些图片,合适以一篇原创的形式写出来。

    依风1986(xurunjiao5339) 发表:这个还没碰到过,TANG566版主不是找安捷伦维修过吗?维修细节怎么样?我知道安捷伦ICP点火要两次,一般第一次点火后熄灭会提示F1245

0
    +关注 私聊
  • 依风1986

    第5楼2013/08/26

    应助达人

    那就期待您佳作了,是不是跟之前论坛版内鼓风系统清洁差不多,维修一次总共花了多少银子?

    tang566(tang566) 发表:维修细节有拍些图片,合适以一篇原创的形式写出来。

0
    +关注 私聊
  • tang566

    第6楼2013/08/27

    应助达人

    没有更换部件,花费了两千多。

    依风1986(xurunjiao5339) 发表:那就期待您佳作了,是不是跟之前论坛版内鼓风系统清洁差不多,维修一次总共花了多少银子?

0
    +关注 私聊
  • 依风1986

    第7楼2013/08/27

    应助达人

    呵呵,不便宜啊!

    tang566(tang566) 发表:没有更换部件,花费了两千多。

0
    +关注 私聊
  • chenwr

    第8楼2013/08/27

    PE的切割气问题遇到过

0
    +关注 私聊
  • 秋月芙蓉

    第9楼2013/08/27

    的确是个原创的好题材

    千层峰(jxyan) 发表:建议整理一下,参加原创。
    仪器还是要好好保养护理。出问题的耗材该换就换。

0
    +关注 私聊
  • 秋月芙蓉

    第10楼2013/08/27

    这个频率有点高了,之前有版友写了篇这个方面的原创

    chenwr(chenwr) 发表:PE的切割气问题遇到过

0
查看更多