ICP-MS 用EPA的方法测试MDL

ICP-MS

  • 请教各位,我现在需要用Aglient 8800做MDL,参照了EPA的方法。我配制7个相同浓度的样品,测试完成后算MDL。在这个过程中,我加标的浓度是否可以低于我检量线的最小浓度,是否需要先做仪器的IDL,测试结果是否需要减去空白?谢谢!

    Step

    Action

    1



    Estimate the expected MDL.

    2



    Calculate the desired spike level such that MDL is between 10 to 100% of the spike.

    3



    Spike at least 7 replicate samples with analytes of interest. Complete the sample preparation steps for each replicate.

    4



    Calibrate the measurement tool. Measure the analytes. This is done within 1 day.

    5



    Calculate the standard deviation of the 7 measurements. (s)

    6



    Determine the correct t-statistic tdf, 99
    Determine degrees of freedom (df=n-1)
    Find the one-tailed t-statistic for 99% confidence level
    · t-stat is 3.143 for 7 measurements.

    7



    Calculate MDL = t statistic x standard deviation x dilution factor (if not already included in the data).

    8



    Verify: 10% Spike level £ MDL £ Spike level

    9



    Validate that the calculated MDL makes sense.
    +关注 私聊
  • poursuivre

    第1楼2017/03/19

    加标浓度介于2-8倍估计MDL。
    你疑惑的是扣什么空白?

0