【讨论】热电DSQ诊断时老出现ROD DC test failed,如何解决?

气质联用(GCMS)

  • 热电DSQ诊断时老出现ROD DC test failed:
    A Rod DC 50 51.7 passed
    B Rod DC 50 50.7 passed
    Average DC 50 51.2 Failed
    Rod DC A/B Switch passed

    热电工程师说不影响做样。个人觉得肯定会影响离子能量,但不知具体有什么影响?

    DSQ在线帮助中这样说:

    Average DC (V)

    Displays the readback value for the average between DC Drivers. This

    value should be within +/- 0.5 V of 50 V. This represents the

    quadrupole offset that determines the average ion energy. If this

    readback fails by a consistent small amount, i.e. 1 V, the instrument

    will still function. You may have to adjust the ion energy to

    compensate.最后一句补偿离子能量,如何操作?

    离子源洗过的,调谐可以通过,目前调谐参数如下:
    Lens 1: -25
    Lens 2: -7.3
    Lens 3: -25
    Prefilter Offset: -5.4
    Electron Lesn: 15
    Electron Energy: -70
    Emission Current: 100
    +关注 私聊
  • cwjzyh

    第1楼2009/09/02

    试着调过调谐液下面一个特别小的螺丝,有点儿变化但就是通不过,请问大家还有什么高招?

0
    +关注 私聊
  • dickwang2008

    第2楼2009/09/03

    楼主还是问下thermo的工程师,这种硬件问题如果没有遇见过,很难解释!希望楼主问题解决后能将解决方法写出来,大家分享一下。