lzutrunks 2010/07/30
除了考虑蒸发瓶的体积,加热温度和冷凝管温度外,还需要分别考虑是哪一种溶剂在被蒸发 然后根据溶剂本身的蒸发热进行计算 Q = flow rate x heat of vaporization x density Q = heat load in Watt flow rate: in ml/hour heat of vaporization: in Kcal/g density: in g/ml Example Collecting 300 ml of methylene chloride in 30 minutes 1. Convert volume to weight: 300 ml x density of 1.33 g/ml = 399 g 2. Convert minutes to hours: 399 g in 30 min = 798 g/hr 3. Multiply by the latent heat of vaporization: 798 g/hr x 89 cal/g = 71022 cal/hr 4. Divide by 1000 to get Kcal: 71022 / 1.000 = 71,022 kcal/hr 5. Multiply by 1.16 to get Watts: 71,022 x 1.16 = 82 Watt Add a security factor of at least 20% for loss of cooling capacity due to high ambient, long tubing, etc.: 82 x 1.2 = 98 Watt 举个例子,我们在计算的时候基本按照这个方法来做 为了保证可操作性,一般在计算所得数据上再乘1.1到1.2的系数,就万无一失了
lzutrunks
第3楼2010/07/30
然后根据溶剂本身的蒸发热进行计算
density: in g/ml
Example