【求助】请问赛默飞世尔的光电直读可以测试N元素吗?

  • 硫酸里的小虫虫
    2010/09/25
  • 私聊

直读光谱

  • 请问赛默飞世尔的光电直读可以测试N元素吗?
    +关注 私聊
  • mlb2003

    第1楼2010/09/25

    这个厂家的直读没有用过,不过现在直读光谱测试N元素应该不存在问题了!

    硫酸里的小虫虫(haiyang_zou) 发表:请问赛默飞世尔的光电直读可以测试N元素吗?

0
    +关注 私聊
  • wybwyb

    第2楼2010/09/25

    当然没有问题,一个注意事项,你的高纯氩品质必须保证,否则麻烦点,经常校低标和N高标标准化样。

0
    +关注 私聊
  • &超&

    第3楼2010/09/25

    现在的光谱仪器测N应该没问题了,关键你要选一个好的厂家。

0
    +关注 私聊
  • papasea

    第4楼2010/09/25

    首先呢看成分是多少
    不锈钢里面 N含量0.01以上,做出来精度还可以。
    中低合金钢里面 N含量0.01以下,比较难

    制样 磨样 氩气(含N!!)等都很重要。
    如果只要一个参考值现在直读做都没问题,想做很精确必须用ON了

0
    +关注 私聊
  • dyz

    第5楼2010/09/25

    我们这里的4460全都带N通道,不过都是远紫外波长的,Fe基中低合金钢标样,>20ppm的数据完全可信,5-20ppm的数据对操作、氩气纯度和仪器维护要求很严格,一般座位参考。
    转一下操作要求:
    ===========
    D.1 Microalloyed Steels - CNOPS
    Note: CNOPS is the abbreviation of Carbon – Nitrogen – Oxygen – Phosphorus – Sulphur.
    D.1.1 General Consideration
    The analyses of microalloyed steels (CNOPS) require the special VUV option hardware. Therefore you must
    also consult the manual dedicated to this option (VUV Line Analysis) for the handling and maintenance of the
    VUV hardware.
    D.1.2 Working Constraints
    The analyses of microalloyed steels (CNOPS) need the use of absolutely clean stainless steel argon pipes,
    with irreproachable sealing of all connections and fittings.
    For the analysis of oxygen, milling is mandatory because polishing and grinding are contaminating the sam-
    ple surface by adding a supplementary oxide layer. For carbon and nitrogen, the results are usually satisfac-
    tory with grinding, but are nevertheless better by milling. Milling is necessary for the analysis of ultra low
    carbon concentration (ULC). The pure samples used for calibration and standardisation must also be milled.
    The surface quality is important too, the sample must be perfectly flat and the abrasive paper used in case of
    polishing must be is a good state.
    The analysis of the oxygen imposes the hard metal disk to be glued on the analytical table. This constraint
    makes the daily maintenance more difficult, because the table must be removed for the cleaning of the
    sparking chamber.
    The analysis of the carbon, nitrogen and oxygen at very low concentrations is very sensitive to the accumu-
    lation of dust in the sparking chamber, in argon exhaust pipes and to the clogging of the entrance of the out-
    put argon filter. The reproducibility of these elements becomes sensibly degraded if the routine maintenance
    is not done very regularly.
    Moreover for the oxygen, the analysis spots must not crossover: black halos around the sparking area must in
    no way be superposed. After the run of a badly positioned sample (sample not flat or analysis hole not com-
    pletely recovered), iron oxide powder is blown on part or the entire sample surface. This powder affects the
    oxygen results: the sample needs to be surfaced again.
    The content of impurities depends on the sparking positioning between the centre of the sample (inhomoge-
    neous by definition) and the periphery sensitive to micro-leaks. The best position for the sample corresponds
    to a sparking zone so that the exterior of the black halos are situated at around 3-4 mm from the sample edge.
    D.1.3 Particular Procedures
    For the cleaning procedures of the components of the stand, the ordinary procedures (see Routine
    Maintenance chapter) can be followed. However they must be applied f more often or the reasons explained
    in the previous section Working Constraints.
    D.1.3.1 Before Every Run
    ♦ Electrode, top of the table and hard metal disk
    Clean the electrode with the small steel brush (yellow) before every analysis. Then clean the table surface
    and the disk with a tissue paper in order to remove all the dust as well as fingertips.
    If the time interval between two analyses is long (5 minutes or more), the oxygen value is higher on the first
    run because of the stand contamination during the idle time. This first run must be eliminated from the
    average.
    D.1.3.2 Partial Maintenance
    ♦ Hard metal disk, insulator, sparking chamber
    Once a day is a strict minimum. In case of shift work, it must be done at each shift change at least, or every
    400 runs.
    If the disk is not glued, you can also remove it quickly and aspirate the dust accumulated into the insulator. If
    the disk is glued, the table must absolutely be removed in order to aspirate the dust inside the sparking
    chamber. Remove the table with care, so as not breaking the insulator.
    ♦ Argon exhaust, pipes, filter entrance
    At the same time you clean the chamber, the exhaust argon pipe must also be cleaned and the end at the oil
    filter entrance side cleaned out. To do that, the pipe must be released from the filter entrance. Use a cotton
    pad or a pipe cleaner for cleaning into the elbow fitting. Then aspirate wit a vacuum cleaner at the pipe end
    and inside the elbow fitting. Put back the pipe into the fitting.
    After a partial maintenance, 15 runs are necessary to decontaminate the stand from the oxygen
    lead in during the handlings.
    D.1.3.3 Full Maintenance
    ♦ Inside, table, supporting block, cylinder
    Once a week is a strict minimum. In case of high sample throughput or if a large amount of dust is generated
    the rate must be increased. If the instrument operates 24 hours a day, this maintenance may become daily.
    The analytical program must be completely standardised after every full maintenance because of P and S
    elements.
    After a full maintenance, 50 runs are necessary to decontaminate the stand from the oxygen lead
    in during the handlings.
    A 5 l/min flush during 1 to 2 hours before starting analyses helps to recover quicker the
    performances on oxygen.
    D.1.3.4 Argon Output Filter
    The ordinary maintenance of the VUV output filtering device must be done at least once a week.
    D.1.3.5 Lens Cleaning
    Consult the VUV Line Analysis manual for the handling and maintenance of the lens.
    For CNOPS lines, the lens contamination is not significant as long as the instrument is standardised every
    eight hours.
    The problem of the lens appears only after the maintenance of the lens, the highest drift is observed just after
    that maintenance.
    D.1.3.6 Standardisation
    For all the reasons developed in this chapter a daily standardisation is at least required. In case of shift
    working, a standardisation at the beginning of each shift change should be done. However the other required
    maintenance operations must be executed before the standardisation.
    D.1.3.7 Calibration
    If you intend to calibrate yourself new microalloyed programs, it is essential to proceed first to a full mainte-
    nance including cleaning of the inside of the stand, the pipes, the output filter and the lens.
    After that maintenance, run analyses during at least one hour before starting the calibration.
    D.1.3.8 Other Subjects
    For all the other maintenance subjects not explicitly described in this appendix, refer to the Routine
    Maintenance chapter and its Summary Table section.

0
  • 该帖子已被版主-mlb2003加2积分,加2经验;加分理由:鼓励回帖
    +关注 私聊
  • dyz

    第6楼2010/09/25

    想修改重新排版一下,竟然说有非法字符~~~~~~~~

0
    +关注 私聊
  • mlb2003

    第7楼2010/09/26

    说的不错,在M10和LECO氧氮仪上做过对比,N含量10ppm左右的数据基本吻合!

    dyz(dyz) 发表:我们这里的4460全都带N通道,不过都是远紫外波长的,Fe基中低合金钢标样,>20ppm的数据完全可信,5-20ppm的数据对操作、氩气纯度和仪器维护要求很严格,一般座位参考。

0
    +关注 私聊
  • 187174209

    第8楼2010/12/03

    有好几款光谱仪都可以做啊

0
    +关注 私聊
  • thomasxia2003

    第9楼2010/12/03

    做N没有问题

0
    +关注 私聊
  • 木牛谦谦

    第10楼2010/12/03

    可以的,其实THERMO是收购的人家的光谱,应该叫ARL。
    ARL的优势就在于低浓度和低远紫外元素,例如氧啊、氮啊。

0