某日,收到M家药物研究所waters2998检测器报修,反应检测器基线噪声较大。
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100855195605_9200_5509470_3.jpg!w357x195.jpg)
现场查看,氘灯四个月前更换过,目前仪器噪声大约在0.1-0.2mAu,进入控制台,查看氘灯使用时间为1500h,用甲醇冲洗流通池,读取能量,选择波长230nm,值为4698 counts,初步判断为光强度过低导致噪声大。
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100855563946_2165_5509470_3.jpg!w601x309.jpg)
打开仪器外壳,将光路整体取出,清理狭缝:
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100856236582_3296_5509470_3.jpg!w300x450.jpg)
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100856315638_1601_5509470_3.jpg!w300x450.jpg)
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100856403111_9912_5509470_3.jpg!w300x450.jpg)
清理完成后,将狭缝重新安装好,光路装回仪器,开机,用甲醇冲洗流通池,读取能量,选择波长230nm,值为10395 counts,虽有改善,但依旧偏低。
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100858073018_4058_5509470_3.jpg!w603x329.jpg)
关机,拆开光路部分的M1和M2,有明显的光斑和腐蚀的痕迹:
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100858445414_1368_5509470_3.jpg!w600x510.jpg)
使用特定的镜片清洁剂处理后,效果比较明显:
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100858576056_5338_5509470_3.jpg!w603x329.jpg)
用甲醇冲洗流通池,读取能量,选择波长230nm,值为37577 counts。
![](https://ng1.17img.cn/bbsfiles/images/2023/11/202311100859090779_6954_5509470_3.jpg!w690x318.jpg)
甲醇冲洗,噪声测试,基本小于0.05mAu。
该检测器的使用年限大致4年左右,但光路已经腐蚀比较明显,可能与实验室环境(如温湿度、灰尘等)有一定关联。