PPB级元素在聚乙烯瓶中稳定性
原子吸收光谱(AAS)
  • 与这个相关的资料大家可以看看ICP论坛的准确测定系列
    Part-Per-Billion Stability Study
    Experimentation at the ppb Concentration Level in LDPE Bottles
    The purpose of this study was to determine the stability of elements at the ppb level in low density polyethylene (LDPE) bottles. The stability of metals at the ppb level in this container material was of consequential concern. In work reported previously, it was found that LDPE is the cleanest container material (see The Purity and Cleaning of Plastics for details) for trace metals solutions.
    Experimental Design
    A blend of 65 elements from Inorganic Ventures / IV Labs' CMS-SET was prepared at the 0, 2, 10, and 100 ppb concentration level in 1 % (v/v) HNO3 at the start of the study.
    The set consists of the following;
    CMS-1 - 10 礸/mL Ce, Dy, Er, Eu, Gd, Ho, La, Lu, Nd, Pr, Sm, Sc, Tb, Th, Tm, U, Yb, Y in 3.5 % HNO3
    CMS-2 - 10 礸/mL Au, Ir, Pd, Pt, Re, Rh, Ru, and Te in 3.5 % HCl
    CMS-3 - 10 礸/mL Ge, Hf, Mo, Nb, Ta, Sn, Ti, W, and Zr in 3.5 % HNO3 tr. HF
    CMS-4 - 10 礸/mL Sb, As, Ba, Be, Bi, B, Cd, Ga, In, Pb, Se, Tl, and V in 3.5 % HNO3
    CMS-5 - 10 礸/mL Ag, Al, Ca, Cs, Cr+3, Co, Cu, Fe, Li, Mg, Mn, Ni, K, Rb, Na, Sr, and Zn in 3.5 % HNO3
    Only LDPE bottles (500 mL) were used.
    The LDPE bottles were acid leached with 1% nitric acid for 59 hours at 60 °C. New blends prepared in the same way were compared to the original preparation at 1, 3, 25, 75, 137, 300, and 375 days.
    The New blends were compared to the original blend using ICP-MS and the relative % loss was calculated.
    The ICP-MS used is in a clean room limiting environmental contamination (opened 1 % nitric acid solutions in similar LDPE bottles placed around the auto-sampler yielded no detectable environmental contamination at times of ~ 100 hours).
    Measurements of each blend were made in the same LDPE bottle i.e. the blend was not exposed to any other container during the study.
    Results
    Hg was not stable long enough to measure (minutes).
    Au was the next most unstable element, showing instability at the 2, 20, and 100 ppb levels at 3 days.
    Ag showed instability only at the 10 and 100 ppb levels at 137 days.
    Pd showed instability only at the 2 and 10 ppb levels at 3 days.
    Pt and Ta showed instability only at the 2 and 10 ppb levels at 137 days.
    Ir showed instability only at the 2 ppb level at 300 days.
    Mo, Sn, and Hf showed instability only at the 2 ppb level at 375 days.
    All other elements showed no instability at 2-100 ppb for 375 days, including:
    Ce, Dy, Er, Eu, Gd, Ho, La, Lu, Nd, Pr, Sm, Sc, Tb, Th, Tm, U, Yb, Y, Re, Rh, Ru, Te, Ge, Nb, Ti, W, Zr, Sb, As, Ba, Be, Bi, B, Cd, Ga, In, Pb, Se, Tl, V, Al, Ca, Cs, Cr+3, Co, Cu, Fe, Li, Mg, Mn, Ni, K, Rb, Na, Sr, and Zn.
    Observations
    The 1 % nitric acid solutions of the alkali, alkaline, and rare earth elements do not show any instability at the 2-100 ppb level in LDPE.
    The majority of elements studied were found to be stable for 1 year at the 2-100 ppb level.
    Silver (Ag) is the only unstable element found that is stable at the 2 ppb level. Ag's instability is most likely linked to its chloride chemistry (photo-reduction, precipitation).
    Gold (Au) and Mercury (Hg), which are similar in stability, are the most unstable elements and are the only elements unstable at all of the concentration levels studied. They are also reported to stabilize one another.
    Platinum (Pt), Tantalum (Ta), Molybdenum (Mo), Tin (Sn), Hafnium (Hf), and Iridium (Ir) were originally present as a fluoride or chloride complex.
  • 该帖子已被管理者-设置为精华,下面是奖励记录:
  • 第1楼2005/04/14
    这么好的贴子大家为什么无动于是,我下了
  • 第2楼2005/04/14
    东东是很好,只是阅读起来有点吃力,想请问一下你们的实验室正在使用您说的这种低密度聚乙烯瓶吗?我也想购买一些,您知道在哪里有出售吗?谢谢!
  • 第3楼2005/04/14
    确实是一个不错的帖子,很有用
  • 第4楼2005/04/14
    可以到本网展上看看,很多的
    wqf1617 发表:东东是很好,只是阅读起来有点吃力,想请问一下你们的实验室正在使用您说的这种低密度聚乙烯瓶吗?我也想购买一些,您知道在哪里有出售吗?谢谢!
  • 该帖子已被管理者-设置为精华,下面是奖励记录:加5积分,加5声望
  • 第5楼2006/12/10
    哈哈……问个问题:10 礸/mL 中的礸是什么意思?
  • 第6楼2009/05/01
    是错别字?
    haifeilong 发表:哈哈……问个问题:10 礸/mL 中的礸是什么意思?
  • 第7楼2012/07/04