请教:ICP-MS,ICP-AES与AAS有什么性能比较?/

ICP光谱

  • 请教:ICP-MS,ICP-AES与AAS有什么性能比较?/
    +关注 私聊
  • 第1楼2005/04/17

    ICP-MS与ICP-AES比较检测限要好:ICP-MS、ICP-AES与AAS比较可以进行多元素测量

0
    +关注 私聊
  • 第2楼2005/04/18

    看看这写资料你会明白的,介绍的也很详细的
    http://www.instrument.com.cn/show/download/shtml/007979.shtml
    http://www.instrument.com.cn/show/download/shtml/004888.shtml

0
    +关注 私聊
  • 第3楼2005/04/20

    ICP-MS-1
    April 1994
    the selected ions reach the detector.
    The ICP-MS provides information for each atomic
    mass unit (amu or Daltons; more accurately, the
    ratio of the mass of the ion to its charge is
    displayed, and labelled m/z), in the mass region
    3-250 amu. The isotope information can be used in
    several ways; these include isotope ratio
    measurements, often used for Pb and U which do
    not have a constant natural abundance, and
    analysis of samples having unnatural isotope
    abundances.
    Isotope dilution is a method of spiking the samples
    with a known concentration of a pure isotope to
    obtain a very accurate determination of the
    concentration of the element. A pre-requisite of this
    technique is that the element of interest must have
    more than one isotope.
    Detection Limits
    ICP-MS detection limits are very impressive. Most
    detection limits are in the 1-10 parts per trillion (ppt)
    range for solutions. These are as good as, or better
    than, GFAAS detection limits for most elements in
    pure water. ICP-MS can also detect many elements
    that have very poor GFAAS detection limits.
    ICP-MS has typically two to three orders of
    magnitude better detection limits than ICP-AES,
    which has detection limits for most elements in the
    1-10 parts per billion (ppb) range. Recently, some
    ICP-AES spectrometers have shown impressive
    detection limits in the sub ppb region for selected
    elements in an optimized part of the spectrum for
    clean samples.
    It should be noted, however, that the comment
    above about ICP-MS detection limits is for simple
    solutions having low levels of other dissolved
    material. For detection limits related to
    concentrations in the solid, the advantage for
    ICP-MS can be degraded by up to 50 times,
    because of the poorer dissolved solids capability.
    Some common lighter elements, (e.g. S, Ca, Fe, K,
    Se) have serious interferences in ICP-MS which
    degrade the detection limits considerably.
    ICP-MS, or ICP-AES and AAS?—a comparison
    Geoffrey Tyler
    Varian Australia Pty Ltd
    Mulgrave, Victoria, 3170, Australia
    Introduction
    The attractiveness of the inductively coupled plasma
    atomic emission spectrometer (ICP-AES) has led
    many analysts to ask whether it is wiser to buy an
    ICP-AES or to stay with their trusted atomic
    absorption technique (AAS)1. More recently, a new
    technique, inductively coupled plasma mass
    spectrometry (ICP-MS), has been introduced2.
    The ICP-MS offers at first sight, albeit at higher
    cost, the advantages of ICP-AES and the detection
    limits of graphite furnace atomic absorption
    spectrometry (GFAAS).
    This article will briefly describe these three
    techniques, and point out the important criteria by
    which to judge their applicability to your own
    analytical problems.
    For many people with an ICP-AES background, ICPMS
    is a plasma, with a mass spectrometer as a
    detector. Mass spectroscopists would prefer to
    describe ICP-MS as mass spectrometry with a
    plasma source. Either way, the technique is capable
    of giving isotope information. This information can
    help to overcome many of the “spectral”
    interference problems that can occur in the mass
    spectrometer.
    Basically, the sample compartments and plasma of
    ICP-AES and ICP-MS look similar. In ICP-AES, the
    optical spectrum (approximate range 165-800 nm),
    is viewed and measured, either sequentially or
    simultaneously. The simultaneous ICP-AES is
    faster, but more expensive, than sequential
    ICP-AES.
    The ICP-MS extracts the ions produced in the
    plasma into an interface consisting of a sampler
    cone followed by a skimmer cone. This configuration
    enables the pressure to be reduced differentially
    from atmospheric pressure down to a final pressure
    of between 10-5-10-7 Torr. The ions pass through the
    interface into ion optics, which optimize the ion
    paths. Neutral species are removed from the beam
    at this stage, either by the vacuum pumps or by
    collision with a photon stop. The ions then pass
    through a mass filter, usually a quadrupole, before

0
    +关注 私聊
  • 第4楼2005/04/20

    Interferences
    The three techniques present different types and
    complexities of interference problems. For this
    reason, we will look at each technique separately.
    ICP-MS interferences
    1. Spectral
    The spectral interferences in ICP-MS are
    predictable and number less than 300.
    Polyatomic and isobaric interferences are found
    where a species has a similar mass to the
    analyte, whereby the resolution of the
    spectrometer (generally around 0.8 amu) will
    not resolve it, e.g. 58Ni on 58Fe, 40Ar on
    40Ca,40Ar16O on 56Fe, or 40Ar-Ar on 80Se.
    Element correction equations (similar in
    principle to inter-element correction in
    ICP-AES) can be used. In many cases
    alternative isotopes with lower natural
    abundances may be employed. The use of
    mixed gases (small percentages of other gases
    such as nitrogen, added to the main argon gas)
    can sometimes be effective in reducing
    interferences.
    2. Matrix acids
    It should be especially noted that HCl, HClO4,
    H3PO4 and H2SO4 can cause considerable
    spectral problems. Polyatomic interferences are
    caused by Cl+, P+, S+ ions in conjunction with
    other matrix elements like Ar+, O+, H+.
    Examples are, 35Cl40Ar on 75As and 35Cl16O on
    51V.
    The avoidance of HCl, HClO4, H3PO4 and
    H2SO4 in ICP-MS is paramount for most
    analyses. Where this is not possible, separation
    chromatography (microcolumns) may be used
    before the sample is introduced into the
    plasma. This is a method many favour to get rid
    of the unwanted species, and it also creates an
    opportunity to preconcentrate at the same time.
    Other techniques used to overcome these
    problems are: electrothermal vaporization
    (ETV), and mixed gases. Another very
    expensive alternative is a high resolution
    magnetic sector ICP-MS which can resolve
    masses less than 0.01 amu apart. This enables
    many of the spectral interferences to be
    eliminated.
    Solutions for ICP-MS analysis are normally
    prepared in nitric acid.
    3. Doubly charged ions
    A doubly charged ion will cause a spectral
    interference at half the m/z of the singly
    charged ion, e.g.138Ba++ on 69Ga+ or 208Pb++ on
    104Ru+. These interferences are few and can be
    considerably minimized, or effectively
    eliminated, by optimizing the system before
    proceeding with the analysis.
    4. Matrix effects
    Transport effects include spray chamber effects
    and differences in viscosity between sample
    solutions and calibration standards. This will
    change the efficiency of aerosol production
    from one solution to another. Matrix matching is
    usually required, although internal
    standardization can be used as an alternative
    method. The rapid scanning speed of ICP-MS
    does give superior results when using an
    internal standard.
    5. Ionization
    Ionization effects can be caused by samples
    containing high concentrations of group I and
    group II elements. Matrix matching, sample
    dilution, standard addition, isotope dilution,
    extraction or separation by chromatography
    may be necessary.
    6. Space charge effects
    Space charge effects occur mainly behind the
    skimmer cone, where the net charge density
    becomes significantly different from zero. The
    high ion density leads to interaction between
    ions present in the ion beam causing
    preferential loss of the light ions in the presence
    of heavy ions, e.g. Pb+ on Li+ 3. Matrix
    matching, or careful choice of internal
    standards across the mass range of analytes,
    will help to compensate for these effects,
    although this may prove difficult in practice.
    Isotope dilution will be effective though
    expensive, but the simplest and most effective
    method is to dilute the sample.
    ICP-AES interferences
    1. Spectral
    ICP-AES spectral interferences are more
    numerous and are more difficult to solve. There
    are more than 50,000 ICP-AES spectral lines
    documented, and the matrix can cause
    considerable problems, which makes a high
    resolution spectrometer mandatory for the
    analysis of samples such as steels, chemicals
    and rocks. Inter-element correction, used
    extensively in simultaneous ICP-AES, can have
    only limited success.
    The background in ICP-AES may be elevated
    or structured, requiring an off line background
    correction. Sophisticated dynamic background
    correction, if available, is very useful to improve
    accuracy. Different molecular species such as
    OH give peaks or bands which can cause
    analytical problems at low analyte
    concentrations, degrading the detection limits in
    real samples.
    The background in ICP-MS is so low, typically
    <20 counts/second, that it doesn’t pose a
    problem. This is a major reason for the superior
    detection limits of ICP-MS.

0
    +关注 私聊
  • 第5楼2005/04/20

    2. Matrix effects
    Like ICP-MS, ICP-AES can use internal
    standards to overcome matrix effects such as
    spray chamber effects and viscosity differences
    between samples and calibration standards4.
    3. Ionization
    Interference from easily ionizable elements can
    be minimized by careful choice of individual
    element conditions6 or by adding an ionization
    buffer, i.e. by adding an excess of a group I
    element.
    GFAAS interferences
    The interferences in GFAAS have many sources.
    They include:
    1. Spectral
    There are a few spectral interferences in
    GFAAS when deuterium background correction
    is used, but these can be eliminated by use of
    Zeeman GFAAS.
    2. Background
    For many matrices careful programming of the
    ash stage is required to minimize the
    background signal during the atomization. The
    use of chemical modifiers can be helpful in
    increasing the allowable ash temperature. For
    example, a Ni chemical modifier for Se
    determinations allows ash temperatures of up
    to 1000 °C before Se loss. The use of Zeeman
    background correction can give an
    improvement in accuracy compared with D2 arc
    background correction in many GFAAS
    applications.
    3. Vapor phase interferences
    These can be caused by the atomization of the
    analyte into a cooler gas environment. These
    interferences have been minimized in recent
    years by isothermal tube design and use of
    platforms so that the sample is atomized into a
    hot inert gas environment.
    4. Matrix effects
    Matrix effects are caused by variable retention
    of the analyte on the graphite tube, depending
    on the sample type. The dry and ash stages
    can have a dramatic effect on the shape of the
    transient peak. The use of matrix modifers
    (e.g. PdCl2) and hot injection can be quite
    effective in minimizing these effects; also the
    use of peak area measurement can be
    advantageous in some cases5.
    Ease of use
    For routine analyses, ICP-AES has matured in
    automation to the point where relatively unskilled
    personnel can use methods created by the ICP-AES
    specialist. Until recently, ICP-MS was still the
    domain of the specialist chemist making fine
    adjustments before performing routine analysis.
    The trend to simplicity has been evident since 1993
    and will continue in the future. One of the reasons
    for this is full computer control of parameters stored
    within a method. Another reason is the use of a
    multitasking graphical user interface, to show the
    operator several indicators of data integrity on the
    same screen. The use of such software also has a
    very positive effect on method development time.
    Before this software became available, ICP-MS
    method development was a highly complex and
    time-consuming task. GFAAS, although relatively
    simple for routine analysis, requires considerable
    skill in setting up the methods.
    Total Dissolved Solids (TDS)
    Recent ICP-AES spectrometers have been able to
    analyze routinely up to 10% TDS and even up to
    30% for simple salt solutions. Although the analysis
    of 0.5% TDS for ICP-MS may be possible for a
    limited timescale, most chemists are happier with
    0.2% maximum TDS. This should be borne in mind
    when the original sample is a solid. The ultimate
    detection limit for some elements in ICP-MS may
    not be so impressive when expressed in the solid,
    compared with ICP-AES. GFAAS can cope with
    extremely high levels of dissolved solids.
    Linear Dynamic Range (LDR)
    ICP-MS can have a LDR in excess of 105. Various
    methods for extending the linear range up to 107
    include de-sensitizing one of the ion lenses, use of
    detector analog mode, or use of a separate Faraday
    cup as a second detector. These should be used
    with caution, however, as high matrix component
    concentrations may cause problems best solved by
    dilution. For this reason, and because of the
    problems with high levels of dissolved solids,
    ICP-MS should be mainly the domain of trace/
    ultratrace analysis. The exception to this is when
    using isotope dilution. With this technique very good
    results have been obtained with high concentrations.
    GFAAS has a limited LDR of 102-103. It can be used
    for higher concentrations if a less sensitive line is
    selected.
    Trace to major element analysis may be performed
    by ICP-AES because of its 105 LDR. ICP-AES is
    ideal for analysis up to and including percentage
    levels. For this reason ICP-AES, in addition to
    ICP-MS or GFAAS, is often needed to fulfil
    laboratory requirements.
    Precision
    The short-term precision of ICP-MS is generally
    1-3%. This is improved routinely by use of multiple
    internal standards. The longer term precision (over a
    period of hours) is still <5% RSD. The use of isotope
    dilution can give results of very high precision and
    accuracy, although the cost can be prohibitive for
    routine analysis.
    ICP-AES has generally precisions of 0.3-2% RSD in
    the short term and again less than 5 % RSD over
    several hours.

0
    +关注 私聊
  • 第6楼2005/04/20

    GFAAS, however, will generally have short term
    precisions of 0.5-5% RSD. Longer term precision is
    a function of the number of graphite tube firings,
    rather than time.
    Sample throughput
    ‘The ICP-MS has an incredible capacity to analyze a
    vast number of samples for trace elements. Typical
    analysis time is less than 5 minutes/sample, for the
    whole suite of required trace elements. For some
    applications this may only take a couple of minutes.
    Consulting laboratories find the sample throughput a
    major advantage.
    While the speed of ICP-AES analyses will depend on
    whether simultaneous or sequential instruments are
    used, generally this can vary from 2 to 6 minutes
    per sample. Simultaneous ICP-AES can be faster,
    typically 2 minutes/sample, but sometimes its
    accuracy can be compromised by spectral
    interferences present with some types of samples
    (e.g. rocks).
    The speed of GFAAS is typically 3-4 minutes per
    element per sample (assuming 2 replicates).
    Automated overnight runs can be performed, and
    this will improve throughput of samples.
    Total sample throughput can be a major factor
    favouring ICP-MS in the busy laboratory. The
    following examples (expressed as solution
    concentrations), will give a guide:
    1. One to three elements/sample, at sub/low ppb
    concentration will generally be better by
    GFAAS, assuming the elements of interest can
    be determined by this technique.
    2. 5-20 elements/sample at sub ppm-% levels will
    generally be better by ICP-AES.
    3. 4 or more elements/sample at sub ppb and ppb
    concentrations will generally be better by
    ICP-MS, if the number of samples to be
    analyzed is high.
    Unattended operation
    ICP-MS, ICP-AES and GFAAS can all operate
    unattended overnight, because of the modern
    automated designs and the safety inherent in the
    use of inert argon gas in these techniques. For
    highest productivity, overnight operation is
    mandatory.
    Cost of ownership
    The running cost of ICP-MS is more than ICP-AES
    because several components have a limited lifetime
    and have to be replaced. These include the
    turbomolecular pumps, the sampler and skimmer
    cones and the detector. The torch and nebulizer
    have similar lifetimes for both ICP-AES and ICP-MS
    techniques. If ICP-AES is chosen instead of ICP-MS
    the laboratory will probably require GFAAS as well.
    Hence, the cost of graphite tubes for the latter has
    to be taken into account. In all three techniques the
    cost of argon is a significant budget item, with the
    ICP techniques requiring more than GFAAS.
    Capital cost
    This is always a difficult subject to quantify because
    it will depend on the amount of automation, the
    accessories and the supplier. In very approximate
    terms, you can estimate that an ICP-AES will cost
    twice as much as a GFAAS and 2-3 times less than
    ICP-MS. It should be noted, however, that the
    accessories can distort these figures considerably.
    Another cost that needs to be taken into account is
    that ultra trace analysis requires a clean laboratory
    and ultra pure chemicals. These are not cheap.
    Accessories
    Being a very rapid sequential method, ICP-MS can
    utilize transient signals in multi-element mode. This
    opens the way for a host of accessories.
    Electrothermal vaporization, laser ablation, glow
    discharge and spark ablation can obviate the need
    to dissolve the sample. Some accessories provide
    the means of separating the matrix from the sample
    and/or to pre-concentrate. These include hydride
    generation, and various forms of chromatography
    (e.g. HPLC, ion chromatography, microcolumns).
    The advantage of separation by chromatography for
    speciation work has only been fully realized in
    ICP-MS. This is due to the low concentration levels
    of interest in environmental, toxicological, medical
    and food samples.
    Although ICP-AES can use some of the above
    accessories, their cost and their marginal
    advantages have meant that we have rarely seen
    many of them routinely used.
    Summary
    To advise anyone what to buy is always difficult.
    Look at your present and future needs, and answer
    the check-list questions in table 1. This should help
    you to decide.
    It should always be remembered that no technique
    will satisfy all your requirements. The techniques are
    complementary. There will always be samples where
    one technique is better suited for the analysis than
    another.
    Table 2 shows a simplified comparison of the three
    techniques. Table 3 compares detection limits.

0
    +关注 私聊
  • 第7楼2005/04/20

    References
    1 G.Tyler, AA or ICP - which do you choose?
    Chemistry in Australia, Vol 59, No 4,
    pp 150-152, April 1992.
    2 A.R. Date and A.L. Gray, Applications of ICPMS,
    Blackie, Glasgow, UK, 1989.
    3 K.E. Jarvis, A.L. Gray, and R.S. Houk,
    Handbook of ICP-MS, Blackie, Glasgow, UK,
    1992.
    4 M.Thompson, J.N. Walsh, Handbook of
    Inductively Coupled Plasma Spectroscopy,
    Blackie, Glasgow, UK, 1983.
    5 J.E. Cantle, Atomic Absorption Spectroscopy,
    Elsevier, 1982.
    6 Analytical Methods for Liberty ICP
    Spectrometer, Varian publication 85 100938 00,
    Chapter 5, pp 81-82.
    7 J.Olesik, Elemental Analysis Using ICP-OES
    and ICP-MS, Anal. Chem. Vol 63 No 1, Jan 1
    1991 pp 12A-21A.
    Table 1. Checklist of analytical requirements
    1. How many samples/week?
    2. What are the sample types? (steels, rocks,
    effluents, soils, etc)
    3. What method of dissolution may be employed?
    4. How many and what elements need to be
    determined?
    5. What are the concentration ranges?
    6. What sample volume is typically available?
    7. What other options/accessories are being
    considered? Why?
    8. How important is isotope information to you?
    9. How much money is available to purchase or
    lease costs/month?
    10. What is the cost of ownership and running
    costs for the techniques to fulfil the
    requirements?
    11. What skilled operators are available to you?

0
    +关注 私聊
  • 第8楼2005/04/20

    6
    Table 2. Simplified comparison of ICP-MS, ICP-AES, GFAAS
    ICP-MS ICP-AES Flame AAS GFAAS
    Detection Excellent for Very good for Very good Excellent for
    limits most elements most elements for some some elements
    elements
    Sample all elements 5-30 elements 15 seconds/ 4 mins/element
    throughput 2-6 min/sample /min/sample element/sample /sample
    Linear dynamic 105 105 103 102
    range (108 with range ext'n)
    Precision
    Short term 1-3% 0.3-2% 0.1-1% 1-5%
    Long term (4hrs) <5%* <5%*
    * precision improves with use of internal standards
    Interferences
    Spectral few common almost none few
    Chemical (matrix) moderate almost none many many
    Ionization minimal minimal some minimal
    Mass effects high on low NA NA NA
    Isotopes yes no no no
    Dissolved solids
    (maximum tolerable 0.1-0.4% 2-25% 0.5-3% >20%
    concentration)
    No. of elements >75 >73 >68 >50
    Sample useage low high very high very low
    Semi-quantitative yes yes no no
    analysis
    Isotope analysis yes no no no
    Routine operation easy easy easy easy
    Method skill skill easy skill
    development required required required
    Unattended yes yes no yes
    operation
    Combustible
    gases no no yes no
    Operating cost high high low medium
    Capital cost very high high low medium/high
    Table 3. Detection limit comparison (mg/L)
    Element ICP-MS ICP-AES Flame AAS GFAAS
    As <0.050 <20 <500 <1
    Al <0.010 <3 <50 <0.5
    Ba <0.005 <0.2 <50 <1.5
    Be <0.050 <0.5 <5 <0.05
    Bi <0.005 <20 <100 <1
    Cd <0.010 <3 <5 <0.03
    Ce <0.005 <15 <200000 ND
    Co <0.005 <10 <10 <0.5
    Cr <0.005 <10 <10 <0.15
    Cu <0.010 <5 <5 <0.5
    Gd <0.005 <5 <4000 ND
    Ho <0.005 <1 <80 ND
    In <0.010 <30 <80 <0.5
    La <0.005 <0.05 <4000 ND
    Li <0.020 <1 <5 <0.5
    Mn <0.005 <0.5 <5 <0.06
    Ni <0.005 <10 <20 <0.5
    Pb <0.005 <20 <20 <0.5
    Se <0.10 <50 <1000 <1.0
    Tl <0.010 <30 <40 <1.5
    U <0.010 <30 <100000 ND
    Y <0.005 <0.5 <500 ND
    Zn <0.02 <1.0 <2 <0.01
    ICP-MS, ICP-AES, Flame AAS: Detection limits (defined on the basis of 3 standard deviations of the
    blank)
    GFAAS: Sensitivity (0.0044 absorbance) measured with 20 mL of sample
    ND: Not determined