You should get better HPLC system suitability.
You can not consider only single parameter. Maybe you should consider the follwing parameters at the same time according to USP requirments:
--Capacity factor =(T2-T0)/T0 should be more than 1.0 but less than 10, The better range is 2 - 7;
--Tailing factor=( 5% Peak Width)/2f not more than (NMT) 2.0;
--Column Efficiency N =16* Square of (Retention time of analyte / peak width measured at baseline of the extrapolated sides to baseline) = L/H should be more than 2000 or other specified value;
--Resolution R between two peaks should be more than 2.0