tomdreams
第3楼2007/09/20
Baseline Correction
Select this option to apply a baseline correction (Baseline) to the sample scan (S) that is collected from the instrument as transmission values.
This correction is thus:
(S)/(Baseline)
Zero/Baseline Correction
Select this option to apply a 100%T baseline correction (100%TBaseline) and a zero line correction (0%TBaseline) to the sample scan (S). This correction is performed with the raw transmission data and is thus:
((S)-(0%TBaseline))/(100%TBaseline-0%TBaseline)
With this option, the Cary will prompt you to perform a 100%TBaseline first, followed by a 0%TBaseline when you perform a baseline collection.
This baseline option corrects for any inherent variations in the electronic zero line of the instrument. You should use this baseline option if your samples have areas of very low transmission (high absorbance) as any variations in the instrument's zero line will effect your measurements if they you do not correct for these. You should refer to ASTM method E903 for further details.
Zero SRA correction Cary 100/300/400/500/500i/4000/5000/6000"no"Cary 50
Select this option to apply a 100 %T baseline correction (100%TBaseline) and a zero line correction (0%TBaseline) to a sample (S) measured with an Absolute VW Specular Reflectance Accessory (SRA).
Note
The VN SRA and the single bounce relative SRA accessories should not use this correction. Select a Baseline/Zero correction instead.
This correction is performed with the raw transmission data and, due to the double bounce on the sample, is thus:
Sqrt(((S)-(0%TBaseline))/(100%TBaseline-0%TBaseline))
With this option, the Cary will prompt you to perform a 100%TBaseline first, followed by a 0%TBaseline when you perform a baseline collection. You should refer to ASTM method E903 for further details.
Zero *Std. Ref. Correction
Select this option to apply a 100 %T baseline (100%TBaseline) correction and a zero line correction (0%TBaseline) to a sample (S) measured with a reflectance accessory which is not absolute (e.g. a Diffuse Reflectance Accessory or the Variable Angle Specular Reflectance accessory). The Cary then multiplies the result by the appropriate value in the 'Baseline Stdref' file which are the transmission values associated with the Standard Reference material you used to establish the baseline. Refer to the help for the Diffuse Relectance accessories for further details about Standard Reference materials.
((S)-(0%TBaseline))/(100%TBaseline-0%TBaseline)) x (Baseline Stdref)
Note
If you select this option the Baseline button in the main window will not be enabled until you have retrieved a Std Ref file. Click this button to browse for and select a Std Ref file. This file must be in the ASCII *.CSV file format.
tomdreams
第4楼2007/09/20
可以使用Cary Help,上面就是。
我感觉:
1.Baseline Correction
100%T基线校准
2.Zero/Baseline Correction
0%T和100%(以空气为参比)两种情况下校准
3.Zero SRA correction
Cary 50没有此项功能,因为Cary 50是单波长的仪器。
4.Zero *Std. Ref. Correction
0%T基线校准和100%T标准参比溶液校准。
Cary Help上面还有这些的别的说明,目录如下:
Software Application
Scan
Software User Interface
Buttons
Setup
Baseline
Cary Help 里面有详细的公式,你自己可以参考。