1 进入Chromeleon的Wellness 面版,在External Conductivity Cell CALIBRATION 中点击 INSTRUCTION,了解操作过程.
2 将泵的出口与电导池的入口用至少1000PSI的反压管连接.
3 在EXTERNAL CONDUCTIVITY CELL CALIBRATION 中点击CELL35C的命令,等待电导池升温稳定至少5MIN.
4 以1.0ML/MIN 的流速泵送1MM KCL 校正溶液.
5 等待电导值稳定后(至少15MIN)执行CALIBRATE 命令
6 电导值应为147正负2us.
7 点击LOG,储存新的校正值.
8 用去离子水清洗电导池至电导读数接近于零后停泵.
9 将管路恢复原状.
回答完毕,给钱吧,呵呵祝好.
jellyhu2000
第2楼2009/03/14
Calibrating the Conductivity Cell
When to Calibrate
• After every 6 months of use
NOTE Do not use this procedure to calibrate a new cell.
1. Open the ICS-1000 Wellness Panel in Chromeleon
2. On the Wellness Panel under Electric Conductivity Cell
Calibration, click Offset Cal.
3. When the offset calibration is complete, click Slope Cal on the
Wellness Panel.
4. Disconnect the pump output line from the injection valve.
5. Disconnect the line from the suppressor ELUENT OUT port to the cell
inlet and connect the pump output line directly to the cell inlet.
6. Verify that there is a minimum of 7 MPa (1000 psi) of backpressure.
7. On the Wellness Panel under Conductivity Cell Calibration, click
Cell 35 C. Allow the cell to reach this temperature and then wait an
additional 5 minutes to let it stabilize.
8. Select 1.00 mL/min to begin pumping 1.0 mM KCl through the cell at
1.0 mL/min.
9. Wait until the total conductivity reading stabilizes (in about
15 minutes) and then select Calibrate on the Wellness Panel.
After calibration, the conductivity reading should be 147.00 ± 2 μS
and the cell constant should be between 120 and 180. If this is not the case, contact Dionex for help.
10. Click Log to record the new calibration values in the Audit Trail.
11. Flush the KCl solution from the system by pumping deionized water
through the cell. When the conductivity drops to less than 1 μS/cm,
stop the pump.
12. Reconnect the pump to the injection valve and reconnect the line
from the suppressor to the cell inlet.
Items Needed Description
1.0 mM KCl solution ---- Prepare by dissolving 0.07456 g of reagentgrade
KCl in 1 liter of 18-megohm DI water.
Backpressure tubing to provide at least 7 MPa (1000 psi)-----Use 0.076-mm (0.003-in) ID yellow PEEK tubing (P/N 049715).
Thanks !But chinese is easy to read !
iso2008beijing
第6楼2010/12/01
1 进入Chromeleon的Wellness 面版,在External Conductivity Cell CALIBRATION 中点击 INSTRUCTION,了解操作过程.
2 将泵的出口与电导池的入口用至少1000PSI的反压管连接.
3 在EXTERNAL CONDUCTIVITY CELL CALIBRATION 中点击CELL35C的命令,等待电导池升温稳定至少5MIN.
4 以1.0ML/MIN 的流速泵送1MM KCL 校正溶液.
5 等待电导值稳定后(至少15MIN)执行CALIBRATE 命令
6 电导值应为147正负2us.
7 点击LOG,储存新的校正值.
8 用去离子水清洗电导池至电导读数接近于零后停泵.
9 将管路恢复原状.