给大家看看碰撞池是怎么消除干扰的

ICP-MS

  • 是动画不能作为附件添加,咋办?
    请给指教,我这里有两个动画关于碰撞池的,想给大家看看
    +关注 私聊
  • 第1楼2005/04/22

    ORS can be operated without using highly reactive gases. Octopole design allows higher cell pressures increasing reaction cross section due to better ion beam focussing
    He Mode
    Collisionally Induced Dissociation – the interference is broken apart when it impacts the He atom
    Energy Discrimination – the larger polyatomic species (greater ionic radii) loose more energy than the smaller analyte species and cannot enter the QP – effectively a molecular filter
    Aided by the Agilent ShieldTorch System – all ions enter the Octopole with well-defined narrow energy distribution
    H2 Mode
    Charge Transfer – electron transferred from (or to) hydrogen, producing a neutral species

    Actually both modes use Energy Discrimination, at least in part